Thread: KH Director reading too high - tried a lot of the known fixes

Hybrid View

Previous Post Previous Post   Next Post Next Post
  1. #1
    Join Date
    07.05.2020
    Posts
    28

    Default

    Vinny posted this
    Quote Originally Posted by Vinny View Post
    Hello Oscar,

    The correct tolerance when using the KHD and Reference solution is +/- 0.4° dkH (0.2 from KHD and 0.2 from Reference solution).
    Given your results, you are within the tolerance range.

    If you would like to get the KHD result a bit closer to the 7.5 mark, check again your tube lengths and calibrate your pump heads. When calibrating pump heads, you can use a digital scale to get the most precise measurement; 1g = 1mL

    You can also verify your water sample tube volume by temporarily disconnecting the tube going into the KHD and manually filling both ends of the tube. Empty this tube into a digital scale or measuring cup and type-in the amount into the tube volume field in the KHD settings page.
    In addition to putting the filter at the end of the sample tube I found tiewrapping the filter in a 25mm pvc tee helped the consistency (and kept the filter where I wanted it)

  2. #2
    Join Date
    17.03.2021
    Posts
    9

    Default

    Quote Originally Posted by Bob_The_Reefer View Post
    Vinny posted this

    In addition to putting the filter at the end of the sample tube I found tiewrapping the filter in a 25mm pvc tee helped the consistency (and kept the filter where I wanted it)
    Thanks, I'm happy with it now I've determined the calibration volumes by weight and results are as expected - less happy that I had to purchase microscales and I've used almost an entire bottle of reference fluid trying to work out what was wrong - in my opinion the documentation should reflect the importance of the calibration volume particularly the reagent and that the volume should be determined by weight

  3. #3
    Join Date
    07.05.2020
    Posts
    28

    Default

    I also ordered a set of coffee scales to improve the calibration process for the KHD and the dosers .... I guess it might also come in handy once my IonD arrives