Results 1 to 10 of 10

Thread: 50mS/cm calibration solution

Hybrid View

Previous Post Previous Post   Next Post Next Post
  1. #1
    Join Date
    11.05.2011
    Location
    Bucharest
    Posts
    26

    Default

    Yeap, you are right. I was comparing the calibration solution using another
    method, that is refractometer, and obviosly the solution is having different properties
    in terms of refracting the light compared to seawater. Two solutions with the same
    conductivity can have different readings to light refraction. I am using digital refractomer
    as a backup unit to double check the salinity of the water, especially when I am preparing
    water for change.

    One more question, maybe you know, what cell constant is having the GHL platinum probe?

  2. #2
    Join Date
    20.02.2009
    Location
    Christchurch
    Posts
    410

    Default

    Quote Originally Posted by kirucd View Post
    One more question, maybe you know, what cell constant is having the GHL platinum probe?
    Sorry, I am not sure what you are asking here?

    The platinum probes are used as they are more resilent to sea water and organic contamination. (But they still can get organic growth on them, so wipe with a cloth or paper towel every week or so)
    They do not seem to be adversely affected by calcium buildup etc.
    (I had a carbon probe on my old Aquadine and this was a real pain as the calcium affected it all the time)
    Call me Paul !
    Note: I am not GHL staff!

  3. #3
    Join Date
    11.05.2011
    Location
    Bucharest
    Posts
    26

    Default

    Every conductivity probe is having a cell constantant, for probes used to measure saturated
    solutions is usually 10 or 5, for brackish water is 1, for low conductivity solutions is 0,1 and so on.
    This is the construction cell constant and based on it the probe can measure conductivity in an interval.
    When you make the calibration of the probe you practically (the software and hardware is doing that)
    calculate the real cell constant of that probe.

    Normally the cell constant is K=d/a where d is distance between electrodes and a is area of electrodes
    and is measured in 1/cm. The usual probes are having this constant 10, or 1 or 0,1 etc. The probe from
    GHL is not having an usual constant and I wanted to know what constant is having.

  4. #4
    Join Date
    20.02.2009
    Location
    Christchurch
    Posts
    410

    Default

    Interesting, I found some info on calculating the constant however it is a little beyond my current knowledge.
    Also not really in my to do list at the moment.

    GHL will have to answer this one!
    Call me Paul !
    Note: I am not GHL staff!

  5. #5
    Join Date
    11.05.2011
    Location
    Bucharest
    Posts
    26

  6. #6
    Join Date
    20.02.2009
    Location
    Christchurch
    Posts
    410

    Default

    Yeah Ok. That is a bit more readable than the info I found last night.
    So to find cell constant you just need to have a feeler gauge, and a vernier caliper.
    Not really sure why you would want this as the profilux is programmed with this data and would be irrelevant to the average user.

    But anyway, measurements are (Platinum GHL probe)
    Width of one electrode .3cm
    Height of one electrode .5cm
    Distance between electrodes .15cm

    Cell constant K=d/a
    K = cell constant (cm-1)
    a = effective area of the electrodes (cm2)
    d = distance between the electrodes (cm)

    .5*.3=.15*2=.30 mm2 = .3 cm2

    .15/.3=.5

    Cell Constant is .5 ?


    Or if you take into account the surface area of only 1 electrode this would be a cell constant of 1!



    I see page 20 refers to more than the physical geometry determining cell constant.
    Call me Paul !
    Note: I am not GHL staff!

  7. #7
    Join Date
    11.05.2011
    Location
    Bucharest
    Posts
    26

    Default

    I wanted to know cell constant in order to know what are the measuring limits of the probe.
    The real cell constant is actually determined by controller in the moment of calibration.
    The physical cell constant that you calculated is only a theoretical one. I don't think the
    cell constant for GHL platinum probe is .5 because you cannot measure seawater with
    a probe that is having this constant. Probes with cell constant around 1 are used for
    brackish water, with 0.1 for pure water, and usually with 10 for solution with high conductivity
    like seawater.

    I was curious also to find out what solution is GHL solution of 50mS/cm because if I am
    going to use for calibration a different solution but with the same conductivity at 25 degrees
    it is possible to have a bad calibration if I am doing the calibration at a different temperature
    then 25 degrees.

    Hope somebody from GHL will tell us what type of probe is the one they are selling.

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •