there is a simple test to check if the measurement is correct:
1. do a precise calibration - fluid bottle for some hours in tank, use the tank water temp. as reference during calibration
2.leave the bottle in the tank: put the probe into the fluid (which has tank temp.) and check if it shows 50.0mS (+/-0.2 is OK and normal)
of course the tank temp. probe must be used for compensation