did I understand correctly - swapping the probes sho that both probes are OK but always the same sensor port will not do anything?

this would mean that the port died (for whatever reason) - but: how is it posible then that you were able to calibrate the probe?

what value does ProfiLux show if youput the redox probe into the calibration fluid? is 220mV displayed? or no reaction here, too?