Page 2 of 2 FirstFirst 12
Results 11 to 17 of 17

Thread: No Conductivity Calibration Option on P3ex

  1. #11
    Join Date
    28.05.2011
    Posts
    15

    Default

    Thanks Gunther I will look into these options and get back to you.

  2. #12
    Join Date
    28.05.2011
    Posts
    15

    Default

    Ok so I hooked up my 2ex and was able to calibrate the conductivity probe no problem except the reading where all wacked. I then did an update on the firmware on the 3ex, calibration option returned. I then tried to calibrate the same way I did it on the 2ex, probe did not read anything when it was in the calibration solution. Once calibration was complete I exited the conductivity probe setting and went back to the main screen. Then I went back into the conductivity probe setting and the calibration option was gone once again.

    What now, if I didnt need the profilux controller to run my lights it would be gone.

  3. #13
    Join Date
    06.02.2012
    Posts
    2,020

    Default

    Sounds like a loose connection or a broken cable. Calibration is not possible if the probe isn´t recognized.

  4. #14
    Join Date
    28.05.2011
    Posts
    15

    Default

    Looks like it is a problem with the board that does Cond/Redox/Level 3, took the board out of my 2ex and put it in the 3ex and it calibration is there. So that means I need to get a new board some how.

  5. #15
    Join Date
    06.02.2012
    Posts
    2,020

    Default

    If you mean the ex-extension, you have to recognize it first before you can use the additional inputs, for more informations see http://support.aquariumcomputer.com/...oad.php?did=75

  6. #16
    Join Date
    28.05.2011
    Posts
    15

    Default

    The EX-extension is recognized on both controllers, it is definitely the card because I now have my 3ex fully running, conductivity probe is reading EXACTLY what my refract reads. One card works perfect in both the other card has a problem and will not allow calibration of the conductivity probe.

  7. #17
    Join Date
    19.02.2009
    Posts
    2,976

    Default

    Please contact your retailer for checking the card.