I do not have the option of calibration of my conductivity probe, running firmware 5.17. How do I correct this?
I do not have the option of calibration of my conductivity probe, running firmware 5.17. How do I correct this?
You can only calibrate through the control panel on the Profilux not in PLC.
Strange ... I just did it on my 3ex last week.
Sensor setting / Conduct / Calibration
Last edited by reeferqq; 10.01.2014 at 05:34.
When I go under conduct there is one spot that is blank.
this menu does not appear, if:
- sensor disabled
- sensor broken or not present
No support or warranty issues over PM! Please send PMs to the moderators only if you have general problems with using the forum! Thanks for helping us to keep the support efficient.
Kein Support oder Reklamationsabwicklung über PM! Bitte senden Sie an die Moderatoren nur PMs bei allgemeinen Problemen mit der Verwendung des Forums! Danke dass Sie uns dabei helfen, den Support effektiv zu gestalten.
Probe is activated so maybe the probe is hooped?? I will try the connect it to my 2ex and see if it works on it.
OK so conductivity probe is in fine working order, hooked it up to my 2ex & was able to calibrate it just fine. So what is the deal with my 3ex?! Need some real tech support here NOW please.
There are tree posibillities you can check:So what is the deal with my 3ex?
- you have connectet the probe to the wrong socket (Redox instead of Conductivity)
- your PL3 is not an ex aka the ex extension isn´t activated (you can check that in main menu of PLC or at status web page
- your PL3 ex extension isn´t mounted correctly. Download the installing manual here and ensure it´s mounted correctly
If all this doesn´t help, it seems the input is defective. Connect your dealer i that case.