Thread: Problem with GHL calibration fluid 50mS

  1. #1
    Join Date
    22.10.2010
    Location
    Netherlands
    Posts
    224

    Default Problem with GHL calibration fluid 50mS

    Dear, GHL

    I recently purchased 2 bottles of your 50mS(at 25C) calibration fluid(Chnr: 112212L, 11/2014).
    When I compare the values with the AquaMedic and Sera calibration fluids, also 50mS at 25C then I notice a difference of 3,8mS.

    The calibration fluids where all the same temperature, but this does not even matter because the measurements have been done with electronics temperature correction. I would expect to find 3 measurements which are the same, The AquaMedic and the Sera values are the same, the GHL is different. The AquaMedic and Sera fluids are also new, fresh from the supplier, so no old or half full leftovers.

    All and all the real readings of my aquarium are better(when compared to my refractometer and aerometer) when I calibrate with the Sera or AquaMedic then when I calibrate with the GHL fluid.

    Am I doing something wrong or is there a faulty batch of GHL 50mS in circulation

    Please investigate and respond

    With Kind Regards,
    Jack

    (Sorry,my mistake this topic should be transfered to the Englisch section)


  2. #2
    Join Date
    18.02.2009
    Location
    Kaiserslautern
    Posts
    7,710

    Default

    I don't know if and what goes wrong with your calibration and/or measurement but I don't think we have a faulty batch, we receive the fluids from a laboratory in Germany, each batch comes with certifications and sample measurement are made

    it is your decision if you trust other's companies fluids more than GHL's products, I don't think we can solve this here
    No support or warranty issues over PM! Please send PMs to the moderators only if you have general problems with using the forum! Thanks for helping us to keep the support efficient.
    Kein Support oder Reklamationsabwicklung über PM! Bitte senden Sie an die Moderatoren nur PMs bei allgemeinen Problemen mit der Verwendung des Forums! Danke dass Sie uns dabei helfen, den Support effektiv zu gestalten.

  3. #3
    Join Date
    22.10.2010
    Location
    Netherlands
    Posts
    224

    Default

    Hi Matthias,

    It is not a matter of trust, it is just that it is strange that the GHL value is different than the other two brands(which is not necessarily a bad thing, because the other two brands can also be wrong). The thing that worries me is that when I calibrate with the other two calibrationfluids the aquarium measurements are closer to the comparisonmessurments that I took with my refractometer and earometer(floater type).
    The only thing and I am looking for is a way to explain this.

    And things can go wrong sometimes, we are all human. For instants there is a batch of AquaMedic 50mS fluid, but what is in the bottle is 100mS. There first reaction was also, "our 50mS fluid is perfect" but al in al it was not.

    Perhaps there is a other way to check the calibrationfluids, which I have not thought at?, or is the only way to send it to a laboratory?


    Kind Regards,
    Jack

  4. #4
    Join Date
    18.02.2009
    Location
    Kaiserslautern
    Posts
    7,710

    Default

    so, meanwhile I personally took 15 bottles out of our stock and tested them

    all of them had perfect readings -> I really doubt that there is a problem with our fluids
    No support or warranty issues over PM! Please send PMs to the moderators only if you have general problems with using the forum! Thanks for helping us to keep the support efficient.
    Kein Support oder Reklamationsabwicklung über PM! Bitte senden Sie an die Moderatoren nur PMs bei allgemeinen Problemen mit der Verwendung des Forums! Danke dass Sie uns dabei helfen, den Support effektiv zu gestalten.

  5. #5
    Join Date
    22.10.2010
    Location
    Netherlands
    Posts
    224

    Default

    Hi Matthias,

    If I may ask.....
    How did test the 15 bottles calibrationfluids, what did you use as reference..... or did you check them against each other and then they all had the same value. Was there also a bottle from the batch I had?

    Greeting,
    Jack




    Quote Originally Posted by Matthias View Post
    so, meanwhile I personally took 15 bottles out of our stock and tested them

    all of them had perfect readings -> I really doubt that there is a problem with our fluids

  6. #6
    Join Date
    18.02.2009
    Location
    Kaiserslautern
    Posts
    7,710

    Default

    we use a calibrated resistor-meter and do the tests with the fluid @ 25°C
    No support or warranty issues over PM! Please send PMs to the moderators only if you have general problems with using the forum! Thanks for helping us to keep the support efficient.
    Kein Support oder Reklamationsabwicklung über PM! Bitte senden Sie an die Moderatoren nur PMs bei allgemeinen Problemen mit der Verwendung des Forums! Danke dass Sie uns dabei helfen, den Support effektiv zu gestalten.