    Hello!

    I have a problem with the PLM-ADIN extension card I bought.
    I have a ProfiLux 3 eX and the firmware is 5.11.

    What I wanted to do is connect an irradiance sensor (W/m2) for a photobioreactor.
    The sensor has an analogical exit of 0-10 V.

    So, I installed the PLM-ADIN and I connected the sensor exit to the D-subminiature (DE-9) in the pins 6 (Analog input 1) and 8 (GND).

    The PLN-ADIN is found in the Profilux and give me the two Voltage values. But the values are always zero.
    The sensor works I tested with a Voltmeter and the volt increase following the warm or the light.

    Any clues?
    Tank you very much in advance!

  2. #2
    Join Date
    06.02.2012
    Posts
    2,020

    Default

    You can test if the card is working by connecting Pin #1 to pin #6 - you should get the 10V.
    If you insert a resistor in the wire, the value should decrease (20 kOhm ~5V; 1 MOhm ~0,2V ....)

    Maybe the possible load of your signal is too low for the ADIN card. You can test that by connecting a 20 kOhm resistor between the GND and the 0-10V output of your sensor (aka create a short-circuit with a 20 kOhm resistor). If the voltage between the connections drop down, the signal load is too low for the ADIN card

  3. #3
    Join Date
    08.01.2013
    Posts
    11

    Default

    I tried to connect the pins 1 and 6 and the measured voltage was 10 V.
    I also inserted a resistor of 270 Ohm (the only I have) in the wire and the voltage decreased to 9,94 V.

    I haven't understand exactly what do you mean with the second paragraph:
    what is the difference between the first circuit with the resistor to test if the card is working?

    Something like this? (R= 270 Ohm)
    2013-01-29 10.43.43.jpg Voltage 1 = 9,94 V
    1.jpeg Voltage 1 = 10,00 V
    2013-01-29 11.20.48.jpg Voltage 1 = 7,30 V

    Anyhow, if the problem would be the low load of signal, can this be solved in some way?

    Thank you very much!
    Last edited by camcam85; 29.01.2013 at 11:59.

  4. #4
    Join Date
    06.02.2012
    Posts
    2,020

    Default

    I tried to connect the pins 1 and 6 and the measured voltage was 10 V.
    I also inserted a resistor in the wire and the voltage decreased.
    So the ADIN card itself is working properly

    I haven't understand exactly what do you mean with the second paragraph
    Maybe a little sketch will make it more clear:130129adin.png

    Measure the sensor voltage without the ADIN card (the 20 kOhm resistor simulates the internal resistance of the ADIN). If the voltage drops down with resistor, the sensor can´t be used with the ADIN the card directly.

    if the problem would be the low load of signal, can this be solved in some way?
    That´s not so easy. You need some kind of isolation amplifier with 1:1 rate

  5. #5
    Join Date
    08.01.2013
    Posts
    11

    Default

    Ok I did what you told me without the ADIN card but actually it was always 0,0 V with or without the resistor.
    So I measured the output of the sensor with the voltmeter and it was 0,0 V. I thought that probably the light of the office wasn't so good for the sensor and I placed on the sunlight and the volts were 2,7. I connect everything again and now works! Sorry for the inconvenience