Quote Originally Posted by doggydoc10 View Post
Yes you did Mathias. I understand. Thank you very much.
But my question is NOT regarding the conduc value that's "!"
At the time of that SMS, level 3 was triggered. I was doing some cleaning and I triggered level 3 while at the same time I was calibrating conduct probe. That's why the conduct value is so off....I understand the alarm going off because of conductivity value, but why if I triggered level 3 to "test it", I did not see the level showing up on the SMS?
Doggydoc
this is possible when the trigger pulse was short enought:

1. ProfiLux (SMS unit) detected alarm
2. when the alarm report was created (happens dynamically) the level state changed again