Thread: Conductivity Probe Calibration Problem

  1. #1
    Join Date
    21.11.2011
    Location
    manchester
    Posts
    108

    Default Conductivity Probe Calibration Problem

    Hi

    I just calibrated my conductivity probe using the 50ms solution.

    All went well but when i put it back in the solution at the same temp as the tank, I got a 49ms reading. I thought that was quite close but decided to calibrate again to see if i could get it more accurate.

    I followed the same procedure again (but it had been in rodi water for about 5 mins) and this time the solution read 32.9ms!

    I did it one more time and it still read 32.9ms. (wish i had left it alone!)

    What have i done wrong? and how can i fix this problem?

    thanks

  2. #2
    Join Date
    19.02.2009
    Posts
    2,976

    Default

    You should simply repeat the calibration procedure.

  3. #3
    Join Date
    21.11.2011
    Location
    manchester
    Posts
    108

    Default

    I have repeated the calibration process 4 times now and the reading of the fluid still comes out at 32.9-32.8

    I have now ran out of calibration fluid.

    could the probe be defective?

  4. #4
    Join Date
    19.02.2009
    Posts
    2,976

    Default

    The first calibration went ok, didn't it?

    Did you change anything before the second try?
    Could you post the ADC (ADC1 / ADC2) values for the conductivity probe reported in ProfiLuxControl?
    If you are using salt water with a ProfiLux3eX or ProfiLuxPlus2eX the range should set to salt water. If ProfiLuxControl shows Conductivity(S), the range setting seems ok.

  5. #5
    Join Date
    21.11.2011
    Location
    manchester
    Posts
    108

    Default

    Yes the first one came out at 49 with the probe in the solution after the calibration

    I left the probe in ro water and tried again

    the ADC values were pretty constant during calibration with ADC1 at 19-20 and ADC2 at 449-451

    The probe is set to sea water in profilux 3ex

  6. #6
    Join Date
    19.02.2009
    Posts
    2,976

    Default

    Hm, the ADC values seems good - the higer value is perhaps a little low - but I have no reverence value now.

    Which firmware version is running and which model do you have?
    Did you a firmware update?

    Maybe a solution:
    save your settings and sensor data via PLC
    make a factory reset on your ProfiLux
    load the settings to the ProfiLux via PLC
    check the ProfiLux settings
    retry the calibration (you could reuse the calibration fluid for those tests - if the calibration went ok you could repeat it with fresh calibration fluid)

  7. #7
    Join Date
    14.10.2009
    Location
    Dubai
    Posts
    168

    Default

    When did you purchase the calibration fluid? as in nov/dec there was a falty batch of mS liquid...
    Forward the barcode of the bottle to GHL for checking!

  8. #8
    Join Date
    18.02.2009
    Location
    Kaiserslautern
    Posts
    7,710

    Default

    I doubt that the calibration fluid is the problem - we had a few bottles with 1.4mS in it instead of 50mS, if this would be the case he would measure complete different values, but these values are nearly OK

    the influence of the temperature compensation is very high for conductivity measurement, that means that the temperature of the calibration fluid must be entered very accurately

    just for testing - to check the probe and hardware with eliminating the temperature influence you can do this:
    calibrate and enter a fix temperature, e.g. 25°C, and set a fixed temperature for the compensation instead of using the temperature probe
    -> after this the ProfiLux should show a reading of 49.8 ... 50.2 mS when immersing the probe into the calibration fluid right after the calibration process
    No support or warranty issues over PM! Please send PMs to the moderators only if you have general problems with using the forum! Thanks for helping us to keep the support efficient.
    Kein Support oder Reklamationsabwicklung über PM! Bitte senden Sie an die Moderatoren nur PMs bei allgemeinen Problemen mit der Verwendung des Forums! Danke dass Sie uns dabei helfen, den Support effektiv zu gestalten.

  9. #9
    Join Date
    19.02.2009
    Location
    MANOTICK
    Posts
    866

    Default

    This thread maybe of some help, not entirely related but could help others with issues and expectations

    http://www.reefcentral.com/forums/sh...light=salinity


    GHL ProfiLux Canada/USA Distribution and support

  10. #10
    Join Date
    11.01.2011
    Posts
    28

    Default

    I'm having the same problem. Readings of 32.8 after calibration.

    sent from the bottom of the sea using tapatalk