Thread: Cond probe does not work after calibration reading, 0 mS only

  1. #1
    Join Date
    05.08.2010
    Location
    Holland
    Posts
    49

    Default Cond probe does not work after calibration reading, 0 mS only

    system: P3ex, sw 5.0.6.0

    reading of prove are
    val 1 0 ms adc1 43,
    val 2 49,5ms adc2 34,

    i've set the nominal value at 49,5 ms (Density 1.021,4)

    Q: why does the probe not view the correct values, now it is 0 Ms. Who can help to solve the issue

  2. #2
    Join Date
    09.07.2010
    Location
    Aschaffenburg
    Posts
    987

    Default

    reading of prove are
    val 1 0 ms adc1 43,
    val 2 49,5ms adc2 34,
    Val 2 should be much highter than val 1

    Q: why does the probe not view the correct values, now it is 0 Ms. Who can help to solve the issue
    Thee are three possiblilities (or a combination of it):
    calibration is done wrong
    a cable is broken
    BNC connector is not connected well

  3. #3
    Join Date
    05.08.2010
    Location
    Holland
    Posts
    49

    Default conductivity probe - problem solved

    i´ve solved my problem, thanks for the tip.