Thread: Defective null plug?

  1. #1
    Join Date
    13.12.2009
    Location
    Alberta, Canada
    Posts
    72

    Default Defective null plug?

    When I am trying to calibrate my redox probe, after I insert the null plug, I get a message telling me that the electrode is defective even though the null plug is inserted.
    Is the PIII referring to the null plug or the probe?
    Could it be a software problem with the expansion box? The expansion box is firmware version 1.03.

    Thanks

  2. #2
    Join Date
    18.02.2009
    Location
    Kaiserslautern
    Posts
    7,710

    Default

    when you inserted the null plug and started the calibration - is the shown ADC value stable? if it is not stable the null plug is possibly defective
    No support or warranty issues over PM! Please send PMs to the moderators only if you have general problems with using the forum! Thanks for helping us to keep the support efficient.
    Kein Support oder Reklamationsabwicklung über PM! Bitte senden Sie an die Moderatoren nur PMs bei allgemeinen Problemen mit der Verwendung des Forums! Danke dass Sie uns dabei helfen, den Support effektiv zu gestalten.

  3. #3
    Join Date
    13.12.2009
    Location
    Alberta, Canada
    Posts
    72

    Default

    ADC value stays stable at 0511.

  4. #4
    Join Date
    18.02.2009
    Location
    Kaiserslautern
    Posts
    7,710

    Default

    then the plug is Ok

    does this message appear when calibrated with plug or with probe at 220mV?
    No support or warranty issues over PM! Please send PMs to the moderators only if you have general problems with using the forum! Thanks for helping us to keep the support efficient.
    Kein Support oder Reklamationsabwicklung über PM! Bitte senden Sie an die Moderatoren nur PMs bei allgemeinen Problemen mit der Verwendung des Forums! Danke dass Sie uns dabei helfen, den Support effektiv zu gestalten.

  5. #5
    Join Date
    13.12.2009
    Location
    Alberta, Canada
    Posts
    72

    Default

    the message appears when calibrated with plug.

  6. #6
    Join Date
    18.02.2009
    Location
    Kaiserslautern
    Posts
    7,710

    Default

    so ProfiLux only calibrated with the null plug and the calibration at 220mV was never started?

    does the shown ADC change if you remove the null plug during the calibration and when you connect the probe instead (the question is - is the redox port "alive"?)

    after what time does the calibration end? is there still time left (see downcounter) or is the time exceeded?
    No support or warranty issues over PM! Please send PMs to the moderators only if you have general problems with using the forum! Thanks for helping us to keep the support efficient.
    Kein Support oder Reklamationsabwicklung über PM! Bitte senden Sie an die Moderatoren nur PMs bei allgemeinen Problemen mit der Verwendung des Forums! Danke dass Sie uns dabei helfen, den Support effektiv zu gestalten.

  7. #7
    Join Date
    13.12.2009
    Location
    Alberta, Canada
    Posts
    72

    Default

    - I initiate calibration sequence
    - I insert null plug when prompted
    - countdown timer starts at 899 seconds and goes down to zero (ADC stable at 0511)
    - at that point I get a message that says "electrode is defective"

    I'll try removing the null plug and replacing with probe during calibration. I'll post those results next.
    Last edited by MitchM; 23.03.2011 at 19:13.

  8. #8
    Join Date
    13.12.2009
    Location
    Alberta, Canada
    Posts
    72

    Default

    Nothing changes when I remove the null plug and replace it with the probe during calibration.
    ADC stays at 0511.

  9. #9
    Join Date
    13.12.2009
    Location
    Alberta, Canada
    Posts
    72

    Default

    I don't know if it is a coincidence or not, but when I am trying to calibrate my salinity probe, I am having similar results.
    Countdown from 899s, ADC stays at 0511 and final message tells me that the electrode is defective.
    Both ports are located on the expansion box.

  10. #10
    Join Date
    07.06.2009
    Posts
    738

    Default

    do you connect the probes to the correct ports ?