Thread: Offset for salinity reading

  1. #1
    Join Date
    19.02.2009
    Location
    MANOTICK
    Posts
    866

    Default Offset for salinity reading

    I tried to recalibrate the salinity proble. It counted down to about 760s on the 50mS calibration before it accepted the value. The cal data reads ADC1 10 and ADC2 512. I am using the "remote control" section to calibrate it, as I cannot find another place in the software to do so.

    I had checked my refractometer against makeup water with which I exactly measured the correct weight to get 1.025. It read 1.024 on my refractometer, 1.017 density with the probe.

    Help!


    GHL ProfiLux Canada/USA Distribution and support

  2. #2
    Join Date
    03.03.2009
    Posts
    20

    Default

    I have forever had a diffrent reading between my newly calibrated refractormeter and my newly calibrated probe. I know a few others also you have to figure out which one you want to follow.
    180g 4'x3'x2'
    2 400w MH
    2 110w VHO
    Profulix top model
    Cond.
    PH
    Redox
    Temp
    Simu-stick

    On it's way
    PTC 6
    "Starfire FM 10 Bulb Dimmable"

  3. #3
    Join Date
    18.02.2009
    Location
    Kaiserslautern
    Posts
    7,710

    Default

    if you want to see readings according to your refractometer then you can enter an offset

    -> this topic has been discussed on several places here, please use the search function