    Default pH calibration question frrom RC

    Dalmerjd writes

    I just got my Profilux +2 and was setting it up. I began the pH probe calibration and ran into a problem. As we know the probe is to be calibrated using the supplied 7 and 9 pH buffer fluid. All was going well until I hit the enter key after setting the 9 pH second calibration point. I got this message:"Calib.Val.Diffe Too Small". I presume this translates into: Calibration Value Difference Too Small. I was then returned to the basic display.

    How do I fix this? Why won't it do this calibration? It will accept a 10 pH as a second point but not a 9. I tried changing the calibration tolerance and the same problem. I tried shutting off the sensor and resetting the hour meter to zero...no difference.

    What do I do?

    I have sent a video clip of what I am doing. I don't think I'm setting both to the same.


    GHL ProfiLux Canada/USA Distribution and support

  2. #2
    Join Date
    20.02.2009
    Location
    Almacelles
    Posts
    150

    Default

    Firmware version of your ProfiLux?
    GHL Iberia

  3. #3
    Join Date
    19.02.2009
    Location
    MANOTICK
    Posts
    866

    Default

    Rectified

    Client moving decimal point.

    i.e instead of scrolling from 10.00 down they were scrolling down using the .00

    This will be an eliminated option in the next software.


    GHL ProfiLux Canada/USA Distribution and support