    Default Temp sensor is needed to be calibarated after Firmware update

    I never got to save or wrote the numbers down. is there any one who knows how to reset the factory setting or numbers?

    thanks!

  2. #2
    Join Date
    18.02.2009
    Location
    Kaiserslautern
    Posts
    7,710

    Default

    you can use the standard calibration values:
    ADC1: 677 and ADC2: 319

    if there is a difference between the reading and the real value you can adjust this with the function "adjust temperature", this modifies the ADCs in order to match the reading

  3. #3
    Join Date
    01.05.2010
    Location
    Lombard
    Posts
    43

    Default

    Thank you, Sir.


    Quote Originally Posted by Matthias View Post
    you can use the standard calibration values:
    ADC1: 677 and ADC2: 319

    if there is a difference between the reading and the real value you can adjust this with the function "adjust temperature", this modifies the ADCs in order to match the reading