Thread: Ph tunze 6101

  1. #1
    Join Date
    31.03.2010
    Location
    The Hague, Netherland
    Posts
    5

    Default Ph tunze 6101

    Hi,
    the Ph reading on the proflux has been reading 9.5 ph, wenn a connect a tunze 6101, wenn a disconnect the tunze form the 230v the reading is 8ph,
    also if a unplug the din plug so the pump end up at full power the ph starts to return to normal.

  2. #2
    Join Date
    19.02.2009
    Posts
    2,976

    Default

    The signal of the pH sensor is very slight.
    Maybe the tunze generate too much electromagnetic noise - and interferes the signal of the pH sensor.

    Try to sepearte both as much as possible - they should not be too close together.

  3. #3
    Join Date
    31.03.2010
    Location
    The Hague, Netherland
    Posts
    5

    Default

    The ph probe is in the sump, i have the power suply from the tunze also connected to a diferent 230v outlet, the ph is then also to high.
    the jump in ph is only when a make a electrical connection white the pomp.

    do jou have a 0-10v liniare optocoupler interface to connect the profilux to the tunze, so it is electrical isolatet.

    sorry for my spelling


    Ruud

  4. #4
    Join Date
    13.03.2009
    Posts
    29

    Default

    the same thing is happening with my radox sensor, 4 tunze 6105s connected to the profilux the reading is 85-150 when the tunze is not connected reading all ok 320.