Thread: Enter calibration values manually after upgrade

  1. #1
    Join Date
    05.04.2010
    Posts
    6

    Default Enter calibration values manually after upgrade

    Hi guys

    I recently had my Profilux II eX upgraded to a Profilux III eX.

    Before the upgrade I wrote all the calibration values for the different probes down.

    Is there any way that I can enter these manually instead of recalibrating all the probes once again?

    BR
    Stefan

  2. #2
    Join Date
    18.02.2009
    Location
    Kaiserslautern
    Posts
    7,710

    Default

    there is a trick:

    1. save all sensor data as a file
    2. then edit this file - you will easily find which values correspondend to the calibration values

    example:
    ...
    10, "CODE_SENSORPARA0ITEM0", 400
    11, "CODE_SENSORPARA0ITEM1", 700
    12, "CODE_SENSORPARA0ITEM2", 101
    13, "CODE_SENSORPARA0ITEM3", 506
    ...
    Items 0 and 1 are the calibration values (here 4.00 and 7.00 for pH)
    Items 2 and 3 are the cal. ADCs

    3. load the modified file back to ProfiLux

  3. #3
    Join Date
    05.04.2010
    Posts
    6

    Default

    Thanks Matthias!

    I will give it a try when I get home.

    BR
    Stefan

  4. #4
    Join Date
    05.04.2010
    Posts
    6

    Default

    Worked like a charm!