Thread: Conductivity probe defective?

  1. #1
    Join Date
    17.03.2009
    Posts
    22

    Default Conductivity probe defective?

    I can no longer calibrate my conductivity probe. I believe there is something wrong with either the probe or the profilux 3. The ADC values are both 18 and the reading is always 99.9 mS. It started about a couple months ago and would read correctly then go to 99.9mS then return to correct reading. The probe does not look physically damaged. Is there a way to test it? How can I tell if its the probe or the profilux? The probe was only used for a little over a year.

  2. #2
    Join Date
    18.02.2009
    Location
    Kaiserslautern
    Posts
    7,710

    Default

    Hi

    both ADC values 18 indicates a broken cable, most likely indside the BNC connector

    are you able to crimp a new BNC connector on it?

  3. #3
    Join Date
    17.03.2009
    Posts
    22

    Default

    I dont have the tools for crimping them.

  4. #4
    Join Date
    17.03.2009
    Posts
    22

    Default

    Anyone in the US have the tools to do this? Is it possible that the probe is also defective or more likely just the Connector?
    I hate to spend another $230 on a Probe that last a year.

  5. #5
    Join Date
    18.02.2009
    Location
    Kaiserslautern
    Posts
    7,710

    Default

    Hi

    this is a standard tool, in the past this tool has been needed to crimp ethernet cables with coaxial cable

    It is possible that something in the connector got loose, it's a try worth.

  6. #6
    Join Date
    19.02.2009
    Location
    MANOTICK
    Posts
    866

    Default

    I am not sure if this is possible matthias can confirm

    Use old BNC connector and cable length from expired pH probe.

    Cut off bad BNC connector and solder on new one, but as you are dealing with millivolts you need to make sure the connection is good.

    I have done this with a pH probe before.


    GHL ProfiLux Canada/USA Distribution and support

  7. #7
    Join Date
    18.02.2009
    Location
    Kaiserslautern
    Posts
    7,710

    Default

    yes, this will work

  8. #8
    Join Date
    16.02.2010
    Posts
    73

    Default

    I have a defective ph probe I can send you if you want to use the wire from it. Where in the US are you?

  9. #9
    Join Date
    18.02.2009
    Location
    Kaiserslautern
    Posts
    7,710

    Default

    please do me a favour - discuss personal things via PM - thanks