Thread: Ex Upgrade

  1. #1
    Join Date
    21.02.2009
    Posts
    35

    Default Ex Upgrade

    Hi GHL,

    Great new site for all enquires. top job

    Question is, I have just upgraded my Pro II into a Pro II EX with the ex upgrade expanstion card. Also at the same time I got in the mail the USB-RS adapter and fitted this and linked the Pro II to my computer. Briliant

    Should it be noticing the new EX upgrade as it still reads as only a Pro II when you go to the main screen that expolanes what version/type you are using.

    If so how do I do this?

    I have downloaded and installed the latest firmware 4.04.

    Roger

  2. #2
    Join Date
    19.02.2009
    Posts
    16

    Default

    You need to complete the final step of the instructions found here: http://www.aquariumcomputer.com/Down...UpgradeKit.pdf

    "As a last step you have to notify ProfiLux that the hardware has changed. For this start ProfiLuxControl II (at least version 3.05)
    and call the menu Extras->Upgrade; the further procedure will be explained there."

    Steve

  3. #3
    Join Date
    21.02.2009
    Posts
    35

    Default

    Quote Originally Posted by UK-GHL View Post
    You need to complete the final step of the instructions found here: http://www.aquariumcomputer.com/Down...UpgradeKit.pdf

    "As a last step you have to notify ProfiLux that the hardware has changed. For this start ProfiLuxControl II (at least version 3.05)
    and call the menu Extras->Upgrade; the further procedure will be explained there."

    Steve
    Thanks Steve works like it should

  4. #4
    Join Date
    21.02.2009
    Posts
    35

    Default

    hi steve,

    Got one mors question regarding the Ex upgrade. There was a part - same as the end of a probe connector, what is this for??

    roger

  5. #5
    Join Date
    19.02.2009
    Location
    MANOTICK
    Posts
    866

    Default

    This is the BNC null plug

    When calibrating the Redox probe you have to give the controller a zero value, so you first fit this to the redox BNC connection on the back of the Profilux and calibrate it with that, then you remove it and calibrate as normal


    GHL ProfiLux Canada/USA Distribution and support

  6. #6
    Join Date
    21.02.2009
    Posts
    35

    Default

    Thanks Michael,

    HA neely toeesd it in the bin. Gotta love the WWW. I will read up on this now.

    Roger

    PS **Note to self - never toss anything out!!!**

  7. #7
    Join Date
    03.03.2009
    Location
    alingsås
    Posts
    133

    Default

    I cant find the menu System-Uppgrade when i set the new EX card
    i use the firmwire 4.0.8

  8. #8
    Join Date
    19.02.2009
    Location
    MANOTICK
    Posts
    866

    Default

    You need matching software i.e 4.08

    Go to Extras at the top and it is in the drop down box


    GHL ProfiLux Canada/USA Distribution and support