Thread: Sensor not locking in place

  1. #1
    Join Date
    09.03.2022
    Posts
    5

    Default Sensor not locking in place

    I've watched a few videos and read the instructions because i wanted to ensure the probe was inserted in the right direction. From what i can see the small notch around the sensor head faces to the right (back of Ion Director) and is on the top side of the probe as you insert it. As recommended in the manual, the small grooves are on the sides to guide the probe inside.
    I insert the probe and screw it tight with 2 hands. The amount of probe sticking out is exactly 22mm as suggested in the literature.
    After priming I "Prepare Measurement Cell". After a few seconds of pumps turning, water shoots out the the sides of the probe and it slides out the machine by about 3cm.

    Have repeated 3 times and keep getting the same result of the probe leaking and ejecting. I'm tightening by hand as much as I can, so I'm not sure what is going wrong that the probe is ejected each time. Any thoughts?
    I'm guessing you dont twist the probe once inside to lock it in place, because I did not read that this was required.
    Thanks

  2. #2
    Join Date
    27.05.2015
    Posts
    2,498

    Default

    Please open a support ticket and include a short video showing what you are referring to when you say the sensor is not locking in. Photos will also help.
    No support or warranty issues over PM!
    Please send PMs to the moderators only if you have general problems with using the forum! Thanks for helping us to keep the support efficient.

    Creating a support ticket
    For support or warranty issues, create a support ticket by clicking here.

  3. #3
    Join Date
    09.03.2022
    Posts
    5

    Default

    Hi Vinny. I've posted a video and photo to the support ticket. Hope it will make sense to you.
    Thanks.

  4. #4
    Join Date
    09.03.2022
    Posts
    5

    Default

    UPDATE: Solution Found
    Raised a ticket and got quick help.
    Turns out somehow I thought to incorrectly connect the waste water line to the 4th pump (similar to KH Director's 3rd pump). A big NO NO as this will build back pressure inside the measuring cylinder and will force the probe out of the holder.

    I fed the waste water tube directly back to the 30 gallon sump and re-ran the process with no leaks this time.
    Thanks for the support.