    Standard KhD is blocked!

    Hi alls!
    On my phone app i have the kh value that my khd has read at 12 oclock.

    I was making some tests some minutes ago, and I started a manual reading of kh.

    I tryed by gcc and it not start.
    I tryed by app and it not start.

    Khd dosos start if i dose manually.

    Pab led color is ok.
    Khd is enabled in system.


    What i have to check?

    Inviato dal mio SM-N975F utilizzando Tapatalk

  2. #2
    Registriert seit
    09.11.2017
    Beiträge
    1.377

    Standard

    Hi,

    please check if the filling level of the reagent is greater than 50ml and that the assigned pumps are the pumps of a connected Doser (check assigned devices and pump numbering).