    Default Fluid remains within the kh at the end of the test

    Is at the end of the kh test left, liquid inside the kh, the dosing pump of the wast does not take out all the liquid

  2. #2
    Join Date
    27.05.2015
    Posts
    2,488

    Default

    A few cm of liquid at the end of the waste water tube is OK. Anything more than that and the pump head should be recalibrated, preferably using a digital scale for best accuracy.