Thema: New Doser, differents values when calibrate

  1. #1
    Registriert seit
    30.05.2018
    Beiträge
    10

    Standard New Doser, differents values when calibrate

    Hi, I have a new Doser 2.1 SLAVE. If I calibrate the pumps 2 or 3 times in a row in fast speed it always gives me different values, for example, the first time 35ml, the second time 35.5, the third time 36.
    This is normal?

    thanks and best regards

  2. #2
    Registriert seit
    27.05.2015
    Beiträge
    2.134

    Standard

    Yes, that is within the acceptable tolerance.
    Take an average of those three results and enter that into the Flow rate field.
    No support or warranty issues over PM!
    Please send PMs to the moderators only if you have general problems with using the forum! Thanks for helping us to keep the support efficient.

    Creating a support ticket
    For support or warranty issues, create a support ticket by clicking here.

  3. #3
    Registriert seit
    30.05.2018
    Beiträge
    10

    Standard

    Ok, thanks