    Default Wrong measurement

    Hello!
    I have one question about measuring liquid in waste container. Kh director constantly reports wronng number acording to real situation. Third pump has been calibrated(38ml/min),sample volume is 10ml,the length of the sample pipe is properly measured,... I dont now what is wrong?
    Thank you , Raff

  2. #2
    Join Date
    27.05.2015
    Posts
    2,491

    Default

    Raff,

    What is wrong about the measurement results?
    • Large difference compared to a conventional test kit? What was the value shown by the KHD and conventional kit?
    • KHD results fluctuate to a large degree with each other? What were the last 3 test results?


    Did you set the pump speeds correctly?
    3-0-3

    Is the vent tube above water?
    This tube should NEVER be submersed in water.

  3. #3
    Join Date
    04.03.2020
    Posts
    3

    Default

    No,no sorry maybe I'm wrong described. Liquid in waste container is let see 500 ml ,waste container of pump 3 measurment is 620ml.

  4. #4
    Join Date
    27.05.2015
    Posts
    2,491

    Default

    If the approximate amount in the waste container is different from what is displayed in GCC or the app, then that means the waste pump should be calibrated.

  5. #5
    Join Date
    04.03.2020
    Posts
    3

    Default

    Thank you very much Vinny.