    Default Weird Sensor problem

    Hello folks

    A couple of days ago I my PH sensor started to act weird.

    From stable readings over the past couple of years ( with a yearly probe replacement ) the reading went from normal reading to sky high.

    Of course I changed to another probe, calibrated and expected the reading tone be normal.

    They were still high and I started to disconnect other stuff as the readings from my aquarium water I a cup was normal but reading from the exact same water in the tank was sky high.

    After I disconnected heaters and sterilizer, the only power source left was the PSU for the mitras lightbar

    So after I unplugged this one the reading went from sky high to normal again.

    The LB PSU and the GHL Profilux 4 is connected to the same wall plug with the same ground reference so I just don't get it.

    Does anyone in here have similar experiences ?

    Does any GHL staffers know why I have this weird issue ?

    I had the impression that the probes was galvanically isolated so this is indeed unexpected.

    For now I will shutoff the PH probe.

    Best regards, Brock