Setup to test once per day (for now) - in 3 days the 3 tests generated 500 ml of waste water.
That seems excessive - am I programmed wrong ?
Setup to test once per day (for now) - in 3 days the 3 tests generated 500 ml of waste water.
That seems excessive - am I programmed wrong ?
Please follow our part 3 KHD setup video for proper setup instructions.
This is likely a result of incorrect pump calibration values.
OK - I'll walk through it again. I would think if it was a calibration issue I would see bad measurements (I'm not) and/or the GHL record of how much fluid is in the container would be off (it's not).
Thanks for your prompt reply - I'll circle back and report here how things go after another calibration.
What sample volume have you set? If you have set 100ml, 500ml after three measurements should be normal because the measurement cell is flushed before each measurement.
No support or warranty issues over PM! Please send PMs to the moderators only if you have general problems with using the forum! Thanks for helping us to keep the support efficient.
Kein Support oder Reklamationsabwicklung über PM! Bitte senden Sie an die Moderatoren nur PMs bei allgemeinen Problemen mit der Verwendung des Forums! Danke dass Sie uns dabei helfen, den Support effektiv zu gestalten.
Currently set to 80ml.
Ok, then 500ml are correct. For flushing the measurement the KH Director uses a little more water than the sample volume. I've just checked that again.
No support or warranty issues over PM! Please send PMs to the moderators only if you have general problems with using the forum! Thanks for helping us to keep the support efficient.
Kein Support oder Reklamationsabwicklung über PM! Bitte senden Sie an die Moderatoren nur PMs bei allgemeinen Problemen mit der Verwendung des Forums! Danke dass Sie uns dabei helfen, den Support effektiv zu gestalten.