Thread: Khd ph probe

  1. #1
    Join Date
    25.05.2019
    Posts
    6

    Default Khd ph probe

    Hi all

    Quick question I’ve noticed that my ph probe reads differently than the that of th ph probe in my p4.
    When measuring the khd probe goes up as high as 7.83 but my p4 ph reading is steady at 8.10.

    Is this correct or is there a variation of .27 when it calculates the ph before working out the Kh reading?

    Thanks

  2. #2
    Join Date
    27.05.2015
    Posts
    2,488

    Default

    Hi,

    The pH probe used in the KHD is not a standard probe.
    It is a modified one used strictly for KH measurement purposes. Being that it is also calibrated with lower pH levels, it will also be the cause for displaying different values than the P4 pH probe.

    KHD probe: Calibrated with 4/ 7 pH
    P4 pH probe: Calibrated with 7 / 9 pH

  3. #3
    Join Date
    25.05.2019
    Posts
    6

    Default

    Thanks for the reply Vinny.

    I’ve found that out thanks are there any suppliers in the UK for the probe?

    I’m not sure what the issue is as after reading it could be dirt in the KHD
    Is this easy to clean and is there any instruction

    Thanks

  4. #4
    Join Date
    25.05.2019
    Posts
    6

    Default

    I’ve ordered a new one from GHL de

    How do you test with the 7.5dkh solution please

  5. #5
    Join Date
    27.05.2015
    Posts
    2,488

    Default

    1. Empty KHD measurement cell
    2. Place water sample tube inside 7.5 reference bottle
    3. Flush measurement cell
    4. Run a KH test