Thread: how to calibrate Conduct probe?

  1. #1
    Join Date
    26.11.2016
    Posts
    4

    Default how to calibrate Conduct probe?

    Just got a Profilux 3 but can't quite find where to calibrate the probes. Don't have a USB plugged in, only wifi.

    Do I need to use a PC and USB direct connection to access the 'calibration' settings?

    I have MyGHL setup on ipad, Expert mode--->Control--->Sensors--->Temp, pH, Conduct, Redox....

    for Temp probe, there is a 'Calibration' function under Maintenance, BUT for all the others, pH, Redox and Conduct(salinity), under Maintenance, there is no 'Calibration'
    Last edited by boshkg; 17.01.2019 at 03:48.

  2. #2
    Join Date
    04.11.2015
    Posts
    1,062

    Default

    You can do the calibration directly on ProfiLux 3 with the LC-Display and navigating to Sensors settings->Cond.(S/F)->Calibration.
    Just 'Activate' the sensor before in the settings.