Thread: Problem calibration the PH probe Ou Director

  1. #1
    Join Date
    31.08.2018
    Posts
    2

    Default Problem calibration the PH probe Ou Director

    Hello,
    I have a problem calibrating the PH probe of my new Kh Director
    after two days of testing no results
    the calibration is very slow then systematic crash of the software with the error message code 28
    could you help me on this subject
    thank you in advance
    good to you
    A.Ledent

  2. #2
    Join Date
    04.11.2015
    Posts
    1,062

    Default

    Most reliable connection type is connecting via USB.
    But you can raise the timeout of the connection in the administration menu to 60000ms. Save, reconnect and retry the calibration.
    No support or warranty issues over PM! Please send PMs to the moderators only if you have general problems with using the forum! Thanks for helping us to keep the support efficient.
    Kein Support oder Reklamationsabwicklung über PM! Bitte senden Sie an die Moderatoren nur PMs bei allgemeinen Problemen mit der Verwendung des Forums! Danke dass Sie uns dabei helfen, den Support effektiv zu gestalten.

  3. #3
    Join Date
    31.08.2018
    Posts
    2

    Default

    Hello,

    thank you for your reply.

    problem solved by USB connection

    best regards.