Thread: KH Director just giving measurement of 4dkh!!!!

  1. #1
    Join Date
    27.01.2017
    Posts
    42

    Default KH Director just giving measurement of 4dkh!!!!

    I’ve had my KH director for over 3 months and now I’ve got a new problem, 2 days ago it gave me a measurement of 4dkh and has continued to give this result in 8 subsequent measurements. My actual value is 7.5dkh as measured by my hanna. I have now flushed the cell, reagent line, emptied the chamber, recalibrated the ph probe and all dosing pumps and the waste pipe is not below the surface. The reagent bottle is full, there are no kinks or blockages in any lines. This is the last thing I need on top of 3 months of trying to get rid of bubbles in the reagent... and not trusting it enough to automate dosing.... any ideas before I take a hammer to this useless thing!!!!!!

  2. #2
    Join Date
    18.02.2009
    Location
    Kaiserslautern
    Posts
    7,710

    Default

    as in FB written:
    If you have 8 measurements which all show the same value: This shows that the KHD works perfectly, it gets reproducable results. If ALL measurements have the same difference to the expected values than it looks very much like a wrong calibration of pumps or probe.

    Please calibrate pH probe (again) and dosers. Post here the settings page of all involved pumps.
    PS: I know you already calibrated, but please do it again, just to be sure.

    And:
    - You use a GHL Doser 2.1 for the KHD and not a Maxi, right?
    - please post a picture of your complete KHD setup where I can see all details, connections, reagent, etc.
    No support or warranty issues over PM! Please send PMs to the moderators only if you have general problems with using the forum! Thanks for helping us to keep the support efficient.
    Kein Support oder Reklamationsabwicklung über PM! Bitte senden Sie an die Moderatoren nur PMs bei allgemeinen Problemen mit der Verwendung des Forums! Danke dass Sie uns dabei helfen, den Support effektiv zu gestalten.

  3. #3
    Join Date
    27.01.2017
    Posts
    42

    Default

    Hi Matthias as said above in my original post I have recalibrated everything, if the level hadn’t dropped from the usually measured 7.4 to 4.0 so quickly I wouldn’t of posted, and as I said in my original post the tanks dkh as measured by my hanna is 7.5dkh, so obviously there is an issue with the director. Since that post Overnight i have had readings of 7.6, 0 and 2.9 dkh so it is all over the place, dkh in a stable SPS doesn’t change by anything like these readings my usual swing is ±0.2 at most. I have flushed everything again and even changed reagent bottles this morning, so I will see if it is still wildly inaccurate when I get home. I will also send this into your support email to look into further.

  4. #4
    Join Date
    27.01.2017
    Posts
    42

    Default

    Hello Matthias
    I left my director measuring throughout the day as stated above, I'm now getting consistant readings of0dkh!!! Is this working perfectly??? ill send in to customer service now.

  5. #5
    Join Date
    25.11.2017
    Location
    Louisville
    Posts
    39

    Default

    Have you checked to make sure none of the dosing heads are sticking? Especially the water input. Hope they get it fixed up for you quick, I know you've had some issues with it before this as well.

  6. #6
    Join Date
    27.01.2017
    Posts
    42

    Default

    Thanks, unfortunately it isn’t that, it’s now showing an error message of Titration took to long and constantly reading 0dkh, I’ve emailed ghl support and I’m sure they will sort it.

  7. #7
    Join Date
    18.02.2009
    Location
    Kaiserslautern
    Posts
    7,710

    Default

    This is under direct care of GHL.

    We will post the news as soon as all has been sorted out.
    No support or warranty issues over PM! Please send PMs to the moderators only if you have general problems with using the forum! Thanks for helping us to keep the support efficient.
    Kein Support oder Reklamationsabwicklung über PM! Bitte senden Sie an die Moderatoren nur PMs bei allgemeinen Problemen mit der Verwendung des Forums! Danke dass Sie uns dabei helfen, den Support effektiv zu gestalten.

  8. #8
    Join Date
    30.03.2018
    Posts
    1

    Default

    Hi, is there any fix for this? I have the exact same problem.

  9. #9
    Join Date
    07.12.2017
    Posts
    6

    Default

    Having the same problem what was the solution to this

  10. #10
    Join Date
    27.01.2017
    Posts
    42

    Default

    It was the internal circulation pump that had failed