Thread: PH Calibration on P3

  1. #1
    Join Date
    02.01.2010
    Posts
    2

    Default PH Calibration on P3

    Hi, after upgrading to P3 I cannot calibrate the PH probe correctly. I have tried calibrating with 4 and 7 and 4 and 9 fluids, but the reading is totally wrong (reading 5.60). The probe(and computer) is under 6 months old, and worked fine before the upgrade. Can you help please?

    Regards,

    Atle.

  2. #2
    Join Date
    19.02.2009
    Location
    MANOTICK
    Posts
    866

    Default

    For marine use you should be calibrating with 7 and 9.


    GHL ProfiLux Canada/USA Distribution and support

  3. #3
    Join Date
    02.01.2010
    Posts
    2

    Default

    Aha, thanks a lot that worked.

    Regards,

    Atle.

  4. #4
    Join Date
    30.12.2009
    Posts
    51

    Default

    i have calibrated my probe but it said use the 4 and 7 solutions??

    do i need to re-calibrate? how do i tell the GHL i'm using 7-9 solutions?

  5. #5
    Join Date
    19.02.2009
    Location
    MANOTICK
    Posts
    866

    Default

    For Marine tanks as you are reading higher than pH 7 you must set the mid range to 7 and the reference to 9.

    How do you tell the controller? By changing the input value when asked to do so.


    GHL ProfiLux Canada/USA Distribution and support