Thread: Micro bubbles forming in reagent tube??

  1. #1
    Join Date
    27.01.2017
    Posts
    42

    Default Micro bubbles forming in reagent tube??

    Hello is anyone else suffering with bubbles in there reagent tube?. My bottle is directly below my dosing pump and the tube is 6” long with no kinks and flows in virtually a straight line to the pump head, After a day or 2 I can see micro bubbles in the tube below the dosing head and I cannot work out how there getting in there, can they be forming somehow in the reagent? It’s frustrating as it gives an inaccurate reading sudden drops in measured value ( ie 5DKH) and so I daren’t let the KHD control the doseing yet.

  2. #2
    Join Date
    06.07.2013
    Posts
    61

    Default

    Quote Originally Posted by Mortie31 View Post
    Hello is anyone else suffering with bubbles in there reagent tube?. My bottle is directly below my dosing pump and the tube is 6” long with no kinks and flows in virtually a straight line to the pump head, After a day or 2 I can see micro bubbles in the tube below the dosing head and I cannot work out how there getting in there, can they be forming somehow in the reagent? It’s frustrating as it gives an inaccurate reading sudden drops in measured value ( ie 5DKH) and so I daren’t let the KHD control the doseing yet.
    Yes I have the odd bubble appear between the dosing head and KH Director . My bottle is directly below too.

    I suspect it maybe sucking air where the tube connects to the doser but it looks a good seal.

  3. #3
    Join Date
    07.04.2017
    Posts
    157

    Default

    I noticed micro bubbles in the reagent fluid itself. Maybe its a temperature issue?

  4. #4
    Join Date
    06.07.2013
    Posts
    61

    Default

    Been reading some forums in the USA, they are having the same problem.

    It does seem to be a issue with bubbles in the reagent bottle.

    https://www.reef2reef.com/threads/kh...289907/page-18

  5. #5
    Join Date
    07.04.2017
    Posts
    157

    Default

    I was thinking maybe using a magnetic stirrer could solve the bubbles in the container. Anyone tried that?

  6. #6
    Join Date
    22.02.2018
    Posts
    7

    Default

    I'm also having the same problem, where the reagent tube from the doser to the kh director gathers air bubbles. This is throwing the kh measurement out by more than 1 dkh <>. Therefore at this time I can not use this to control my kh, as was intended, but only take readings. Which are not always accurate due to the air bubbles.

  7. #7
    Join Date
    06.07.2013
    Posts
    61

    Default

    You won't get a solution from GHL about this problem.
    We are alone with this.

    I found that changing the tube to the link below solved my problem with the bubbles.

    https://www.aquasabi.com/co2/co2-hos...2-m-clear-blue

  8. #8
    Join Date
    18.02.2009
    Location
    Kaiserslautern
    Posts
    7,710

    Default

    Paul, thanks for letting us now.

    GHL is still in the process of finding a reliable solution for ALL users to get rid of the bubbles.

    Please understand we can't just change the tube without examining if this will not cause other side effects.

    We will post here aa soon as we found a working and safe solution.