Reagent Doser fill level not changing
The slave Doser , pump 6 is used to pump reagent to KH Director.
The fill level is still showing 100% after nearly a week of 2 tests per day and several tests during installation.
Screen Shot 2018-02-01 at 13.59.45.jpg
Last edited by paulfars; 01.02.2018 at 15:14.
-
Your having that issue because the incorrect option is selected. Since this pump is being used for the Reagent, you should select CONTAINER IS EMPTIED.
The option you have selected (Container is filled) is used for holding the waste water dispensed by the KHD.
-