Results 1 to 10 of 20

Thread: KH Director not giving me a dkh reading

Hybrid View

Previous Post Previous Post   Next Post Next Post
  1. #1
    Join Date
    28.01.2018
    Posts
    7

    Default KH Director not giving me a dkh reading

    Following on from emails please see screen shots attached
    Firmware.jpg
    KH.jpg

  2. #2
    Join Date
    18.02.2009
    Location
    Kaiserslautern
    Posts
    7,710

    Default

    ok, we see in the screenshot "alarm: range"

    this means the KHD started to measure but it took too long to come to a reasonable measurement value

    when does this happen?

    a) reagent doesn't reach the KHD - check if reagent is pumped into the KHD during measuring and that there is no air pumped

    b) dosing calibration values are incorrect, in this case KHD is not able to calculate correct amounts resp. a correct KH value - check if calibration and settings of all 3 KHD involved pumps are correct
    post screenshots of these 3 pumps here
    No support or warranty issues over PM! Please send PMs to the moderators only if you have general problems with using the forum! Thanks for helping us to keep the support efficient.
    Kein Support oder Reklamationsabwicklung über PM! Bitte senden Sie an die Moderatoren nur PMs bei allgemeinen Problemen mit der Verwendung des Forums! Danke dass Sie uns dabei helfen, den Support effektiv zu gestalten.

  3. #3
    Join Date
    28.01.2018
    Posts
    7

    Default

    Reagent has no air in the tube and is reaching the Doser ok.
    I will calibrate the Doser and try again.
    Thanks

  4. #4
    Join Date
    18.02.2009
    Location
    Kaiserslautern
    Posts
    7,710

    Default

    probe and waste pump shall have about 36 ... 37ml/min at speed 3
    reagent pump approx. 4.5ml/min at speed 0
    No support or warranty issues over PM! Please send PMs to the moderators only if you have general problems with using the forum! Thanks for helping us to keep the support efficient.
    Kein Support oder Reklamationsabwicklung über PM! Bitte senden Sie an die Moderatoren nur PMs bei allgemeinen Problemen mit der Verwendung des Forums! Danke dass Sie uns dabei helfen, den Support effektiv zu gestalten.

  5. #5
    Join Date
    28.01.2018
    Posts
    7

    Default

    Pump 1 is 34ml/min
    Pump 2 is 4.3ml/min
    Pump 3 is 33ml/min

    Still getting 0,0 and alarm error.

  6. #6
    Join Date
    18.02.2009
    Location
    Kaiserslautern
    Posts
    7,710

    Default

    flow seem all a bit low to me, but this is not the reason for this behaviour

    which speeds are selected? please post screenshots of all 3 pumps

    another thought: you did calibrate the pH-electrode, and the calibration was fine at the end? you used fluids pH 4 and pH 7 in the right order?
    No support or warranty issues over PM! Please send PMs to the moderators only if you have general problems with using the forum! Thanks for helping us to keep the support efficient.
    Kein Support oder Reklamationsabwicklung über PM! Bitte senden Sie an die Moderatoren nur PMs bei allgemeinen Problemen mit der Verwendung des Forums! Danke dass Sie uns dabei helfen, den Support effektiv zu gestalten.

  7. #7
    Join Date
    18.02.2009
    Location
    Kaiserslautern
    Posts
    7,710

    Default

    and please deactivate "alarm monitoring" until we found the reason for your issue, maybe there is also another message displayed in the status field
    No support or warranty issues over PM! Please send PMs to the moderators only if you have general problems with using the forum! Thanks for helping us to keep the support efficient.
    Kein Support oder Reklamationsabwicklung über PM! Bitte senden Sie an die Moderatoren nur PMs bei allgemeinen Problemen mit der Verwendung des Forums! Danke dass Sie uns dabei helfen, den Support effektiv zu gestalten.

  8. #8
    Join Date
    10.10.2013
    Posts
    1,000

    Default

    Quote Originally Posted by Paullewis View Post
    Pump 1 is 34ml/min
    Pump 2 is 4.3ml/min
    Pump 3 is 33ml/min

    Still getting 0,0 and alarm error.
    As I also have low values on some pumps, I tell myself that maybe the containers we use to measure the amount of water pump during calibration are not accurate or reliable. Maybe that it would be necessary to provide a reliable and calibratoed containers with Doser2 that guarantees the value of calibrations. As before, KH Director, it was not very important, but with the KH Director, it's very.

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •