    Default Cannot connect 2nd PH probe

    I connected the probe to the Profilux. Unplugged the plugged the Profilux back in but I do not see it reading the second probe. Thank you in advance, Neil

  2. #2
    Join Date
    19.02.2009
    Posts
    2,976

    Default

    Do you have the firmware 4.04 or higher?
    Which ProfiLux type do you have?
    Have you made the corresponding settings after having installed the extension card?

  3. #3
    Join Date
    18.02.2009
    Location
    Kaiserslautern
    Posts
    7,713

    Default

    Don`t forget to enable the second pH-probe under the "extras->display" settings.

  4. #4
    Join Date
    19.02.2009
    Posts
    39

    Default

    Quote Originally Posted by Matthias View Post
    Don`t forget to enable the second pH-probe under the "extras->display" settings.
    Matthias, I tried that and there was no setting for PH probe 2. Also checked under "system", in the software controller, and it only shows sensors 1-4:
    PH-sensor 1
    Temp-sensor 2
    Conductivity- sensor 3
    Redox-sensor 4

    Could it be that the Profilux is not reading the Ph card?
    Neil

  5. #5
    Join Date
    19.02.2009
    Posts
    39

    Default

    oh, i also forgot to mention this has happened with 4.04 and 4.05 firmware. Neil

  6. #6
    Join Date
    18.02.2009
    Location
    Kaiserslautern
    Posts
    7,713

    Default

    are we talking about a ProfiLux Plus II eX?
    If yes - you can`t use slot 0 for measurement cards - this is mentioned in the manual.
    Please use another slot.

  7. #7
    Join Date
    19.02.2009
    Posts
    39

    Default

    Quote Originally Posted by Matthias View Post
    are we talking about a ProfiLux Plus II eX?
    If yes - you can`t use slot 0 for measurement cards - this is mentioned in the manual.
    Please use another slot.
    Matthias, yes it is. I did not see that in the manual but will move to a different slot and try again. thank you, Neil

  8. #8
    Join Date
    19.02.2009
    Posts
    39

    Default

    Matthias, as usual you were correct. Moved card to a different slot and it works great. Thanks again, neil