    Default Calibrating Conductivity probe for freshwater

    I had to select Saltwater then calibrate using the 50mS calibration soultion then change probe setting to (F)reshwater. Is this correct ? If I tried to calibrate it using the 50mS solution with the probe already set to freshwater then the GHL logo flashed red and the display indicated there was an error. Or is there a freshwater calibration solution which would be more suitable ?
    Last edited by Tinpot; 15.02.2017 at 15:40.

  2. #2
    Join Date
    27.05.2015
    Posts
    2,491

    Default

    50mS calibration fluid is for reading saltwater ranges

    1.41mS calibration fluid is for reading freshwater ranges

    In order to use the probe for freshwater purposes, you would first have to switch the probe settings to Freshwater then calibrate with the 1.41mS fluid.

  3. #3
    Join Date
    14.10.2016
    Posts
    31

    Default

    OK. Thanks. But there is no mention of this in the manual.......