Thread: Problem with conductivity psensor connected to the expansion box

  1. #1
    Join Date
    26.11.2016
    Posts
    4

    Default Problem with conductivity sensor connected to the expansion box

    I have a conductivity sensor which is connected to the expansion box 2. It doesn't display any value. Instead displays "***".
    I tried connecting it to the controller(3) itself and-voila- it started displaying a value.
    It shows "***" only when it is connected to the expansion box 2. I tried disconnecting and reconnecting it several times thinking that it would be a connectivity issue but the result was still the same. All the other sensors connected to the expansion box 2 show some value(and deviation when dipped in a different fluid).

    Please Help.
    Last edited by Shaunak; 29.11.2016 at 13:17.

  2. #2
    Join Date
    17.01.2016
    Posts
    23

    Default

    I met the same trouble with you, and it seemed have been solved already. You can try to do this: GCC--"system"--"PAB"--"expansion box 2"--"configuration",change the conductivity sensor from "Saltwater" to "Freshwater" and clink "ok". Then, change the sensor from "Freshwater" to "Saltwater" as the same process.

  3. #3
    Join Date
    27.05.2015
    Posts
    2,488

    Default

    Give the above steps a try. If you haven't done so already, please calibrate all probes connected to the EBX2.