Problem with stirrer
Hi alls,
I'm not able to set my new stirrer. I try several ways but it not works.
I have 2 Doso slave. The first one is a 3 head, for ca,kh amg mg. The second one is a four head and i use itfor kh director. In this way i have the last Doso head(7) Free.
I' m not able to set profilux.
I set stirrer 8, with Doso command. 7 Doso ad command.. 1 cicle/day. 1 minut before command and 45 second ad duration...
Any suggestion?
Thanks
Inviato dal mio SM-N950F utilizzando Tapatalk
-
1 Attachment(s)
Did you choose the right model in the settings of the Stirrer?
Attachment 11407
regards Pit
-
I haven't this column in my gcc.. have you got master or slave units?https://uploads.tapatalk-cdn.com/201...e6721b4d09.jpg
Inviato dal mio SM-N950F utilizzando Tapatalk
-
This evening i'll try to update my profilux.
Inviato dal mio SM-N950F utilizzando Tapatalk
-
I tried to update , Now i'm at 7.15 but i can not choose the stirrer model yet...
Any suggestion?
Inviato dal mio SM-N950F utilizzando Tapatalk
-
Which version does your GHLContolCenter have? Have you tried the expert mode?
Regards Pit
-
The version is 1.1.1.7 date : 19.7.2018
I use Expert mode on gcc
Inviato dal mio SM-N950F utilizzando Tapatalk
-
Please use the Firmware V716 and the GHLControlCenter_V1118
Download
New in V1.1.1.8:
Support for newer magnetic stirrer model (V2) for ProfiLux 3 (with included firmware 6.38), ProfiLux 4 (with included firmware 7.16) and GHL Doser 2/2.1/Maxi Slave/Standalone (with included firmware 1.30)
Sockets/switch channel Operating hour meter can be set for ProfiLux 4
regards Pit
-
Thanks a lot, some minutes ago i find in the forum a specific post. I update the gcc to 1.1.1.8 and then there is the column with model 1 or 2. Thanks a lot. Now it runs[emoji41][emoji41][emoji41]
Inviato dal mio SM-N950F utilizzando Tapatalk