Hello,
My conduct sensor on the expension box giving a alarm.
Even if it is spot on.
And even is Whene the alarm is of.
Thank you
Verstuurd vanaf mijn SM-G950F met Tapatalk
Printable View
Hello,
My conduct sensor on the expension box giving a alarm.
Even if it is spot on.
And even is Whene the alarm is of.
Thank you
Verstuurd vanaf mijn SM-G950F met Tapatalk
Hello,
this could be a communication error reading the probe.
Does the probe work directly on ProfiLux 4?
Hi there,
MatthiasF is asking if you tried this same probe on the conductivitiy port of the ProfiLux controller itself.
This will help us determine if the issue is with the probe or the EXB2.
Can you please post screenshots of your Conductivity probe settings from the EXB2?
What is the firmware version running on both the ProfiLux and EXB2?
Exp 2.04
Profilux 7.16
https://uploads.tapatalk-cdn.com/201...b18d323371.jpghttps://uploads.tapatalk-cdn.com/201...d448198eb1.jpg
Verstuurd vanaf mijn SM-G950F met Tapatalk
He having the same exact issue I am having that I posted about.
It does not matter if I place the sensor off my P4 onto the expansion Box, or use a new salinity probe the expansion box will alarm. His ADC2 values almost match mine.
Attachment 11159
No matter what sensor I place even a new one places the ADC Value over 1000. If I place the one off the expansion box back onto the P4 it shows a ADC value of 462.
This happened after I did the upgrade. Did you do the upgrade and this started to happen?
Yes after the update
Verstuurd vanaf mijn SM-G950F met Tapatalk
Hello,
try setting the measurement range again for the Cond.S probe, which is installed on ExBox2.
Goto System->PAB-ExBox2-configuration and set it to Cond.F and back to Cond.S to apply this setting.
Recalibrate and you should be good.