virtual probe limits ?
Hello Matthias
Is it correct that i can only create one virtual sensor if i have seven (1xtemp, 2xpH, 2xCond, 1xRedox, 1xOxygen) "real" sensors already connected to my Profilux 3 eX (firmware 5.02g BETA) ?
In the manual, under virtual sensors, you have the following Hint:
"In total, 8 sensors can be administrated (sum of actually existing and virtual sensors)."
This is a huge limitation in this fantastic controller! I need at least room for about 8 virtual sensors on top of my existing seven before i can do what i want.
Is it possible to increase the number of virtual sensors in dense configured Profiluxes?
-
this is an issue in the manual, PII has 8 sensors in sum, P3 has 32 sensors in sum
-
Thanks for answer Matthias
This is good and bad news.
The bad is that then there must be a firmware bug at least in firmware 5.02g BETA. I can only create one virtual sensor that behave correctly (on top of my seven real sensors). The second, third, fourth...... virtual sensor i create disappears after power off/on the Profilux 3 eX.
And more strange, before i power off the Profilux 3 eX and the virtual sensors disappear, the second, third, fourth.... newly created virtual sensors only display a static "current value" not mimic the original real sensor, and their control function is not working. The static value displayed is a copy from the real sensor value at the time i created the virtual sensor.
Is this a firmware bug?
-
I must admit I never used so many virtual sensors - I will check this
if it is a bug it will be solved soon