View Full Version : Problem with PH measurment



dimbar
07.02.2011, 08:27
Hello
I have problem with PH measurment, electrode is connected to expantion card
I have colebrated the electrode but in sea water ph is about 10.4, in RO water ph is about 4.8
i tried to change PH electrode but unsuccessfully

Have you any ideas?

Matthias
07.02.2011, 08:28
what are the ADC-calibration values?

dimbar
07.02.2011, 08:31
I will check today evening and let you know

dimbar
09.02.2011, 21:58
I am sorry for delay

PH 7 -ADC:0454
PH 9 -ADC: 0365

Matthias
12.02.2011, 14:18
these values let me assume that the pH-probe has passed its lifetime, you should have a difference of min. 200 ... 300 between 2 calibration

or you used accidentally the calibration fluids in the wrong order?

what pH value is displayed when you immerse the probe into fluid 7.0 and 9.0