View Full Version : Saltwater Conductivity Problems



EgaReef
11.08.2010, 10:53
Hi,

I have a problem with the conductivity probe. The other day the probe began to incorrect values (1002.5).

I clean and calibrate. At that time had a value of 1024.5, but began to decline until 1002.3.

Thinking it was a problem in the probe, I bought a new one. Yesterday, connect the probe and did a new calibration, the probe began to correct values. This morning had a value of 1005.4.

What can be happening?

Ernesto

GHL Support
11.08.2010, 11:44
Hello,

does the temp sensor show correct values? If not, this could perhaps hint to a defective power supply unit.
You can fill some aquarium water in a glass and measure the conductivity. If here the values are correct, then the problem is definitely located in/around your aquarium. Electromagnetic interferences, for example due to a defective pump could then be the reason. In this case remove all electronics one after another until you find the defective device.

EgaReef
11.08.2010, 12:01
Thanks, when I get home I'll try.

boboxx
11.08.2010, 12:11
I had a similar problem a few day ago.. turn out it was a tiny start fish stuck in the middle of the probe fork

EgaReef
11.08.2010, 19:24
Hi,

In calibrating the probe I have a value of ADC 1 of 18 (correct), but the value of ADC 2 is 103 (very low). Also, if you recalibrate the probe get a different value of ADC 2. (76). I remember having a value of 350.

The calibration is made out of the aquarium.

:(

boboxx
11.08.2010, 19:31
hmm ya my values are19 and 424...

try gifing it a good scrub with vinegar (be gental) and try recalibrating with teh proper solution and null plug

EgaReef
11.08.2010, 21:28
hmm ya my values are19 and 424...

try gifing it a good scrub with vinegar (be gental) and try recalibrating with teh proper solution and null plug

I put the probe in vinegar for a few minutes. I have recalibrated and now I have the ADC 1 => 18 and ADC 2 => 427.

I put the probe in the aquarium water and at first had a correct value (1025), but is now 1021.3 and falling in value.

Tomorrow will see the value.

EgaReef
12.08.2010, 08:12
Hi folks,

This morning was 1.024, great news. Values will continue to monitor and comment.

Regards

EgaReef
13.08.2010, 08:04
Bad news,

Yesterday afternoon, the probe value was 1022.4. I calibrate the probe for several hours was correct values, but at night had a value of 1002.3.

Probe Is it wrong?

boboxx
13.08.2010, 11:01
It's not sitting anywhere you ato is filling is it?

EgaReef
13.08.2010, 12:15
It's not sitting anywhere you ato is filling is it?

The probe is in the refuge.

ergo
29.08.2010, 08:44
I have the same problem. Every time I calibrate my probe, after 24 hours it is reading a much lower value. I wonder if there is a problem with the latest software version?

EgaReef
29.08.2010, 17:18
My problem was a pump eheim.

GHL Support
31.08.2010, 08:30
I have the same problem. Every time I calibrate my probe, after 24 hours it is reading a much lower value. I wonder if there is a problem with the latest software version?

Often the reason is a (defective) electronic device that creates an electromagnetic interfering field which disturbs the measurement of the probe. In another case, a customer had put a moor radix into the tank which caused
wrong conductivity values...

Coralleida
01.09.2010, 01:37
Hi Carolin, we found the problem. There was a pump inside the sump which altered the Conductivity probe readings.
After removing that pump, the probe readings are stable.

But I have one question: Why vinegar renews the probe?

GHL Support
01.09.2010, 08:37
But I have one question: Why vinegar renews the probe?
Cleaning the probe can improve the accuracy of the measurement. For an optimal probe cleaning please have a look here:
http://www.aquariumcomputer.com/Forum/showthread.php?1606-Cleaning-Probes

ivolach
01.09.2010, 20:45
I have a similar problem, the conductivity is dropping down. If I shake the sensor in the aquarium water the reading goes back to normal but dropping quickly (from 1.025 to 1.010)

Coralleida
02.09.2010, 00:04
Perhaps it happens to you the same as my friend. Try to unplug devices (specially pumps)