View Full Version : pH calibration question frrom RC



aqua_digital
06.04.2009, 00:52
Dalmerjd writes

I just got my Profilux +2 and was setting it up. I began the pH probe calibration and ran into a problem. As we know the probe is to be calibrated using the supplied 7 and 9 pH buffer fluid. All was going well until I hit the enter key after setting the 9 pH second calibration point. I got this message:"Calib.Val.Diffe Too Small". I presume this translates into: Calibration Value Difference Too Small. I was then returned to the basic display.

How do I fix this? Why won't it do this calibration? It will accept a 10 pH as a second point but not a 9. I tried changing the calibration tolerance and the same problem. I tried shutting off the sensor and resetting the hour meter to zero...no difference.

What do I do?

I have sent a video clip of what I am doing. I don't think I'm setting both to the same.

Coralleida
06.04.2009, 20:27
Firmware version of your ProfiLux?

aqua_digital
07.04.2009, 20:39
Rectified

Client moving decimal point.

i.e instead of scrolling from 10.00 down they were scrolling down using the .00

This will be an eliminated option in the next software.