Hi there
My temp is stuck at 0.0c after an upgrade from 4.08 to 5
The settings import did not work stating there is so serial number in the import
the control program shows my s/n as 1001092047 which is correct
Can you help?
Paul
Matthias
03.03.2010, 11:27
settings can't be imported from 4.xx to 5.xx because the firmware is not compatible
it seem you lost your temp.- calibration data
this can be solved in this way:
- first ensure that the measurement range is correct (pond or tank)
- open the version 4.xx sensorsettings-file with an texteditor (e.g. notepad)
- search for the line with "CODE_SENSORPARA1ITEM2" - the value behind is ADC1
- search for the line with "CODE_SENSORPARA1ITEM2" - the value behind is ADC2
- now enter these 2 values: Click "Calibrate..." in the temp. section of ProfiLuxControl
Sounds simple enough, thanks
Paul
Erm, you've said search for the same value twice
SENSORPARA1ITEM2
SENSORPARA1ITEM2
Do you mean:
SENSORPARA1ITEM1 (notice the 1)
SENSORPARA1ITEM2
Or even 2 and 3?
Couple of screen shots:
http://paulsfishtank.co.uk/photos/log.jpghttp://paulsfishtank.co.uk/photos/temp1.jpg
and here are the settings since upgrade
0, "CODE_SOFTWAREVERSION", 500
2, "CODE_PRODUCTID", 3
6, "CODE_SERIALNUMBER", 1001092047
10, "CODE_SENSORPARA0ITEM0", 700
11, "CODE_SENSORPARA0ITEM1", 900
12, "CODE_SENSORPARA0ITEM2", 430
13, "CODE_SENSORPARA0ITEM3", 198
14, "CODE_SENSORPARA0ITEM4", 0
15, "CODE_SENSORPARA0ITEM5", 800
16, "CODE_SENSORPARA0ITEM6", 30
17, "CODE_SENSORPARA0ITEM7", 0
18, "CODE_SENSORPARA0ITEM8", 0
19, "CODE_SENSORPARA0ITEM9", 1140
20, "CODE_SENSORPARA0ITEM10", 300
21, "CODE_SENSORPARA0ITEM11", 1
22, "CODE_SENSORPARA0ITEM12", 0
23, "CODE_SENSORPARA0ITEM13", 0
24, "CODE_SENSORPARA0ITEM14", 100
26, "CODE_SENSORPARA0ITEM16", 0
27, "CODE_SENSORPARA0ITEM17", 128
28, "CODE_SENSORPARA0ITEM18", 60
29, "CODE_SENSORPARA0ITEM19", 60
30, "CODE_SENSORPARA0ITEM20", 8
31, "CODE_SENSORPARA0ITEM21", 44
32, "CODE_SENSORPARA0ITEM22", 0
33, "CODE_SENSORPARA0ITEM23", 10
34, "CODE_SENSORPARA1ITEM0", 0
35, "CODE_SENSORPARA1ITEM1", 260
36, "CODE_SENSORPARA1ITEM2", 691
37, "CODE_SENSORPARA1ITEM3", 319
38, "CODE_SENSORPARA1ITEM4", 0
39, "CODE_SENSORPARA1ITEM5", 260
40, "CODE_SENSORPARA1ITEM6", 300
41, "CODE_SENSORPARA1ITEM7", 0
42, "CODE_SENSORPARA1ITEM8", 0
43, "CODE_SENSORPARA1ITEM9", 20
44, "CODE_SENSORPARA1ITEM10", 20
45, "CODE_SENSORPARA1ITEM11", 1
46, "CODE_SENSORPARA1ITEM12", 0
47, "CODE_SENSORPARA1ITEM13", 0
48, "CODE_SENSORPARA1ITEM14", 60
50, "CODE_SENSORPARA1ITEM16", 0
51, "CODE_SENSORPARA1ITEM17", 128
52, "CODE_SENSORPARA1ITEM18", 60
53, "CODE_SENSORPARA1ITEM19", 60
54, "CODE_SENSORPARA1ITEM20", 4
55, "CODE_SENSORPARA1ITEM21", 0
56, "CODE_SENSORPARA1ITEM22", 20
57, "CODE_SENSORPARA1ITEM23", 5
10002, "CODE_SENSORPARA0_OHM0", 1207
10010, "CODE_SENSORPARA1_OHM1", 1207
Paul, did you see that the operation is set to off?
Thanks for coming back to me, yes I did see that, it makes no difference whether it's on or off the reading on the screen is still wrong, currently in my tank set to 26c and showing 13c on screen :(
I think the calabration is definatley out, but lines 32 and 33 match 36 and 37 on the new config, most bizarre.
P
Matthias
04.03.2010, 09:08
they are wrong:
34, "CODE_SENSORPARA1ITEM0", 0
35, "CODE_SENSORPARA1ITEM1", 260
please change 34 and 35 to 200 and 300 in the file, than load file back to ProfiLux
Thanks Matthias, that seems to have done the trick, I had to dash to work straight after so I'll check again from home, but it changed from 13c to 27c which is much more likely to be correct.
Paul
Just to close the thread, my temp readings are back to normal.
I'm so glad I made a copy of my settings before I upgraded the firmware :)
Much releived
Paul
Matthias
05.03.2010, 08:27
one question is open for me - why did it happen?
Did you load the 4.xx sensor settings to the 5.xx firmware?
This is not possible because the firmware is not compatible.
In the next version of ProfiLuxControl I will add a protection against this, so it can not happen accidentally in the future.
Yes I did try and load my old settings back, I misunderstood when it said they would have to be entered again.
I have the same problem where is the data?
settings can't be imported from 4.xx to 5.xx because the firmware is not compatible
it seem you lost your temp.- calibration data
this can be solved in this way:
- first ensure that the measurement range is correct (pond or tank)
- open the version 4.xx sensorsettings-file with an texteditor (e.g. notepad)
- search for the line with "CODE_SENSORPARA1ITEM2" - the value behind is ADC1
- search for the line with "CODE_SENSORPARA1ITEM2" - the value behind is ADC2
- now enter these 2 values: Click "Calibrate..." in the temp. section of ProfiLuxControl
Powered by vBulletin® Version 4.2.5 Copyright © 2024 vBulletin Solutions, Inc. All rights reserved.