PDA

View Full Version : Temp Probe reading off



StevieR
02.09.2020, 13:01
I have tried 2 different temp probes and both are way off by as much as 4-5 degrees , I think it may be a setting issue .
How can I fix this issue ?

Vinny
02.09.2020, 16:55
We need more details.

Which ProfiLux are you using?
Where any adjustments made already to the default ADC values?
What are you using as a reference tool?

StevieR
02.09.2020, 17:29
Hi Vinny, thanks for the reply . It's a 3.1n, and I have an aquatronica with 2 temp probes reading 25.5c , the ghl is reading 31.5c at the moment .
ADC1 IS 320
ADC2 is 480

Thilo
03.09.2020, 10:12
This are the ADC values for a digital temp. sensor.
To restore the ADC values of your analogue sensor, please open a ticket (https://www.aquariumcomputer.com/support/tickets-enquiries) and tell us the serial number and the component number of the board (can be found on a sticker on the board).

StevieR
03.09.2020, 14:43
This are the ADC values for a digital temp. sensor.
To restore the ADC values of your analogue sensor, please open a ticket (https://www.aquariumcomputer.com/support/tickets-enquiries) and tell us the serial number and the component number of the board (can be found on a sticker on the board).
Thanks Vinny , will do