View Full Version : Need help with replacing sampled H2O.



arking_mark
18.08.2020, 19:31
I would like to use my 4th doser pump to replace saltwater used by the KDH.

How do I figure out the amount to replace?
The sample size is set to 65ml
The sample tube volume is 33ml
The measured waste is ~85ml
The sample pump draws 110ml and then returns unused portion
The reagent pump seems to use about ~1ml

By my calculations h2o taken from tank should be 110 - 33 = 77. However, based on waste quality it should be 85-1 = 84ml.

What am I missing?

Also, can the doser pull from ~50ft away?

Thilo
21.08.2020, 07:15
Before each measurement the measuring cell is rinsed with 10 ml sample water. So the total amount per measurement is sample volume + 10ml.
To the waste water, the dosed reagent from the rinsing process and the measurement is added. Further deviations are caused by inaccurate calibration and sample tube length.

The doser can dose over several feet. How far exactly also depends on the height. It would be best if you tried it out.

arking_mark
21.08.2020, 13:45
I'm not sure your # is correct. I'll recalibrate the pumps, but I'm pretty sure I have them calibrated well.

Based on your answer, the total water draw is sample size plus 10ml, which would put me at 65 + 10 or 75ml.

This is very close to the actual pump draw which is 110ml minus the tube volume of 33ml which is 77ml. (Close is not good for such an expensive unit)

The fact that there is ~85ml of actual waste water produced is way more troubling to me. ~1ml of reagent is used which means that it actually pulled ~84ml of sample water.

I'll recalibrate and re-measure everything to be certain its set up correctly.

There is one additional thing I noticed. In my setup, the filter hangs at an angle and never completely fills with water. If it volume is completely included, could that account for the extra water. Does the filter need to be completely vertical? If so which way?

Sent from my SM-G988U1 using Tapatalk

Thilo
21.08.2020, 13:54
Yes, that might be the problem. It is best if the tube + filter are installed continuously upwards.
You can also measure the sample volume by filling the tube (and filter) completely and then emptying it into a measuring cup to measure the volume.

arking_mark
21.08.2020, 13:58
I'll fix the filter so its vertical. I'll post my results.

Sent from my SM-G988U1 using Tapatalk

arking_mark
21.08.2020, 18:55
Ok. So here is what I did and my results.

1. I re-ran and re-measured the sample tubing to ensure the filter is verticle. This resulted in 102 inches of tubing. The calculated sample tubing volume with filter is is 40ml.
2. I then just ran the test, which resulted in an initial sample draw of 52ml followed by a waste of 8ml
3. Then the sample pump did a 65ml draw (= to my set sample size) followed by a waste of 75
4. The reagant draw was 1ml

Expected waste would be 12ml + 65ml + 1ml = 78ml actual was 5ml greater than this.

Since these numbers don't add up, I pulled the sample tube off of the KH Director and did a measured dose of 40ml (the calculated volume) and got 5ml dosed into a graduated cylinder. I'm assuming that the 5ml extra is because that calculation is wrong. To be sure the pump was calibrated correctely, I did anoter dose of 12ml which it hit exaclty, so that doesn't seem to be the issue.

This 5ml is exactly the missing volume of waste.

So it would seem that the device pulls 12ml for the initial clean and 65 for the sample volume.

I assume that I just need to correct the sample tube volume to 35ml to get a more precise KH measurement and the expected 77ml.

Is this correct?

arking_mark
21.08.2020, 20:50
So after making the above adjustments, we were still off by 3ml. Expected waste was 79ml and actual waste was 82ml.

So I filled the sample tube and then completely emptied it which resulted in 38ml. I then recalibrated the pump and it went from 37.5ml to 39ml...and reran the test. I'm not sure why the pump calibration resulted in a different number...maybe it's wearing in as I've only had it a week.

Test results coming shortly....

arking_mark
21.08.2020, 21:08
...and finally I got expected waste of 77ml plus ~1-2ml Reagent!! Running it a 3rd time to confirm.

arking_mark
21.08.2020, 21:23
...and 3 waste readings that are perfect.

One other note, with all these tests, the KH only varied from 6.8 to 7.2.

Thilo
24.08.2020, 11:06
Okay, great. Thank you for sharing your findings and solution.

arking_mark
24.08.2020, 16:44
I guess I spoke too soon. This morning I took a KH measure and got 9.2 which was 2 higher than any previous readings. I quickly pulled my other test kit and came back with a 7.5 reading which was in the ball park of the KHD as of midnight. I reran the KHD again and got an 8.3 reading and then remeasured and recalibrated the sample tubing. Then I renan the measurements 3 more times and now am at 8.7. So I'm not sure what is going on. Some differences I'm seeing:
- I am now getting the 10ml in the initial waste and the waste tubing remains full. Before, I was getting only 8ml and the tube was completely empty be the end of the initial waste
- By moving the sample tube to the display, I can now see that the clearing of the tube for a the next sample doesn't completely clear there is about 2ml still in the tubing. If I remove the tubing from the water whil it clears, it all comes out.

Please help. I want to be able to trust these readings...

Something to note, is that this is an AIO tank and doesn't have a sump, so sample tubing runs 6 feet to the tank and goes about 6 inches into the tank. I started dosing Triton which should increase KH, but have seen only about a .2 increase in KH since starting dosing from Saturday. I also did a 10% water change on Sunday which didn't increase KH, it was still at a steady 7.

Thilo
25.08.2020, 14:38
This sounds like a smal amount of fluid is left in the measurement cell because of a problem with the pump calibration and then the measurement is not correct.

arking_mark
25.08.2020, 14:54
How do I fix this?

Thilo
26.08.2020, 13:32
Check the flowrates of the pumps. For further support, it would be best to open a ticket: https://www.aquariumcomputer.com/support/tickets-enquiries/