View Full Version : Leak detected error



ZaneTer
28.02.2018, 10:13
Has anyone had this error before? What causes it? How do we fix it?
Why isn’t there a decent F***ing manual out for this kit?

Vinny
28.02.2018, 11:04
Where is the leak coming from?

The probe locking nut?

If so, please add some Vaseline to the rubber ring as noted in the manual.

ZaneTer
28.02.2018, 11:13
Hi Vinny thanks for the reply but to be clear there is no leak but it is reporting that I have a leak, if it was leaking then it would make sense why I have an alarm but it isn’t leaking

Matthias
28.02.2018, 16:52
This alarm is triggered when there is water detected INSIDE the device (the KHD would not know if there is a leak outside, e.g. in the tubing, how should it)

possible reasons:
- nut was not tight, water came out and got somehow into the device
- leak in internal tubing or measurement cell - not very likely because all devices must pass a pressure test


So there is only one way to find out:
- Unplug the device (no power, no communication)
- open it and check for water

no matter what the result is:
a) no water - false alarm
b) water

you need to contact service@ghl-kl.de in order to get this fixed

ZaneTer
01.03.2018, 06:24
Thank you vinny and this is exactly the response I was hoping for.