Matthias
17.02.2018, 10:23
Very important!
We didn't mention this in the manual (we just didn't think that this will be made wrong), but it happened already a few times and caused wrong measurment results.
The KHD ventilation/overflow tube (from rear connector) must NOT end below the water surface of your waste water - it must be hanging free ABOVE the water surface, it MUST not be possible to suck water into it!
We will add this to the instructions ASAP.
We didn't mention this in the manual (we just didn't think that this will be made wrong), but it happened already a few times and caused wrong measurment results.
The KHD ventilation/overflow tube (from rear connector) must NOT end below the water surface of your waste water - it must be hanging free ABOVE the water surface, it MUST not be possible to suck water into it!
We will add this to the instructions ASAP.