gino555
20.01.2018, 18:23
I am trying to set up my KHD and the KH menu does not appear on the left hand side.

I am running a P3 with firmware 6.31 and the latest version of GCC. I can see the KHD on the PAB under system but dont have the menu on the left hand side for KH:confused:

Matthias
21.01.2018, 10:34
Hello,

as I posted a few days ago I am still working on the firmware update 6.32 for ProfiLux 3. I think it will be done in 1 or 2 days. I will upload the updated ASAP.

Thanks for understanding and your patience.

gino555
23.01.2018, 00:23
Hello,

as I posted a few days ago I am still working on the firmware update 6.32 for ProfiLux 3. I think it will be done in 1 or 2 days. I will upload the updated ASAP.

Thanks for understanding and your patience.

thanks I have now upgraded the firmware but the KHD is not measuring....

i have calibrated the pumps and the probe

put in all the data but when I click start measurement nothing happens :confused: i have a yellow and white flashing logo!

10114

Matthias
23.01.2018, 09:39
your doser has min. firmware 1.26 as the manual requires?

the filling level of the reagent pump is sufficient?

gino555
24.01.2018, 23:05
firmware 1.26 just updated to 1.27, dosing reagent level is at 498ml

I can not vent internal tube, flush cell or empty cell, the only button that works is the calibrate probe button

the logo is flashing yellow/blue

Gael
24.01.2018, 23:46
Hi,

Exactly same for me :-/

Thank

Matthias
25.01.2018, 09:07
what means "does not work"?

buttons are greyed out and can't be pressed or buttons are normal and can be clicked but nothing is happening?

your screenshot is not helpful, the buttons in question are not visible

Gael
25.01.2018, 09:16
Hi,

->buttons are normal and can be clicked but nothing is happening

Thank

Matthias
25.01.2018, 09:17
and the pump numbers are correct?

1,2,3 are the actual numbers of the doser used for KHD?

Gael
25.01.2018, 09:21
I can't answer for gino555, but for me my setup, it's pumps 6,7,8, settings are good and calibration done.

Gael
25.01.2018, 12:55
Do you find the probleme Matthias?

Thilo
25.01.2018, 13:31
The pumps are working when you dose manually from the ProfiLux 3 (not from the Doser) in GHL Control Center?

Gael
25.01.2018, 13:33
The pumps are working when you dose manually from the ProfiLux 3 (not from the Doser) in GHL Control Center?

Yes for me.

gino555
25.01.2018, 17:49
my pumps are 1,2,3 and can dose from the "Dose now" button in GCC. i also calibrated the pumps and seemed to be fine. pump two came out at 6ml flow rate (on slow) for the reagent not sure if that is normal?

Matthias
25.01.2018, 17:56
my pumps are 1,2,3 and can dose from the "Dose now" button in GCC. i also calibrated the pumps and seemed to be fine. pump two came out at 6ml flow rate (on slow) for the reagent not sure if that is normal?

6ml seems a bit high to me, I expect something around 4.5ml

Matthias
25.01.2018, 17:58
just to be sure: your P3 is updated to firmware 6.32 and you use GCC1.1.1.0?

Gael
25.01.2018, 18:09
GCC 1.1.1.0 2018-01-24

***** System report *****
Controller properties
Name: ProfiLux 3.1T eX
Firmwareversion: 6,32
Device address: 2
PAB address: 17
Slot count: 3
Level inputs: 4
Sensor 1: pH-value
Sensor 2: Temperature
Sensor 3: Conduct.(S)
Sensor 4: Redox
Onboard 1-10V outputs: 6
Onboard socket outputs: 8

Slot/card properties
Slot number 0
Name: SERIAL
Class: Communication card

Slot number 1
Name: 2L4S
Class: Intelligent card
Firmwareversion: 1,00
1-10V outputs: 2
Socket outputs: 4

Slot number 2
Name: PWC
Class: Communication card

PAB devices

Name: SMS-Module
Firmwareversion: 3,04
PAB-address: 19

Name: Expansion Box
Firmwareversion: 1,08
PAB-address: 18

Name: VorTech-Controller
Firmwareversion: 1,04
PAB-address: 20

Name: ProfiLux Touch
Firmwareversion: 1,06
PAB-address: 21

Name: GHL Doser 2 Slave
Firmwareversion: 1,27
PAB-address: 22

Name: GHL Doser 2 Slave
Firmwareversion: 1,27
PAB-address: 23

Name: KH Director
Firmwareversion: 1,00
PAB-address: 24

gino555
25.01.2018, 20:04
***** System report *****

Controller properties

Name: ProfiLux 3.1N eX
Firmwareversion: 6.32
Device address: 1
PAB address: 17
Serialnumber: 1210063001
Slot count: 3
Level inputs: 4
Sensor 1: pH-value
Sensor 2: Temperature
Sensor 3: Conduct.(S)
Sensor 4: Redox
Onboard 1-10V outputs: 6
Onboard socket outputs: 8


Slot/card properties

Slot number 0
Name: SERIAL
Class: Communication card

Slot number 1
Name: PWC
Class: Communication card

Slot number 2
none


PAB devices

Name: VorTech-Controller
Firmwareversion: 1.03
PAB-address: 19
Serialnumber: 1008056024

Name: Powerbar 6
Firmwareversion: 1.08
PAB-address: 18
Serialnumber: 1207054072

Name: ProfiLux Touch
Firmwareversion: 1.06
PAB-address: 20
Serialnumber: 1207057055

Name: GHL Doser 2 Slave
Firmwareversion: 1.27
PAB-address: 22
Serialnumber: 1801082017

Name: KH Director
Firmwareversion: 1.00
PAB-address: 21
Serialnumber: 1801088054

gino555
25.01.2018, 20:32
just updated to GCC 1.1.1.0 and it is still the same. buttons change from grey to blue when I click them but no response from the doser or KHD

Gael
26.01.2018, 08:02
What can we do to help GHL to diagnose the problem?
Someone has KH Director successfully working with P3?
Thank

Matthias
26.01.2018, 08:13
Someone has KH Director successfully working with P3?
Yes. Me.

Our 2nd Level support will contact you directly for further support.

Thilo
26.01.2018, 08:30
Could you two please send us your Settings.par file to support@ghl-kl.de. To get the file go to "Backup and restore->Transmit all settings from ProfiLux to file" in GHL Control Center.

Gael
26.01.2018, 08:39
Hi,

Send for me.

Thank

paulfars
26.01.2018, 11:22
I have the same problem as the other two .

Cannot vent internal tube. Cannot flush measurement cell and cannot empty measurement cell.

Im using P3 with 6.32 firmware and using latest version of GCC.

I think I have done everything correctly . I will send my settings to your email also

Thilo
26.01.2018, 15:01
I have checked your settings on two different ProfiLux 3 and the KH Director buttons did work.
Could you please check if GHL Control Center is asking for activating the controls again when connecting and if yes click on 'Yes' or on the display of the P3 if there is a new menu entry for enabling the controls again.
You could also try to reset the PAB settings by clicking "System->System reset" and reassign the PAB devices (maybe try with KH Director and Doser 2 only first).

paulfars
26.01.2018, 15:56
Have reset PAB and just re asigned KH Director and Doser, it made no difference.

I did not see any message on the p3 screen asking for activation

Gael
26.01.2018, 18:18
Reset PAB and just re-asigned KH Director and my two Dosers, same problem.
No message for activation too.

(And had to check settings and calibrate all sensors of expansion box after having re-asigned expansion box, do not do that if you not have time for that after. And also all measurment data lost.... Use reset pab with caution!)

Waiting for months and still can't use 10 days after having received it... so frustrating ^^ :-)

paulfars
26.01.2018, 18:20
If you have backups of settings and sensor data you can reload them after resetting PAB without loosing anything

Gael
26.01.2018, 18:25
Except measurement data.

gino555
26.01.2018, 19:14
Settings sent.


Can I leave the ph probe in the unit? As I have now switched the unit off and removed the probe and replaced the cap.

paulfars
26.01.2018, 19:28
I put my PH probe back in its liquid package just in case the KHDirector was dry

Gael
26.01.2018, 19:30
I put my PH probe back in its liquid package just in case the KHDirector was dry

Same for me

paulfars
26.01.2018, 21:00
Anybody tried the firmware file that was emailed?

Gael
26.01.2018, 21:03
Anybody tried the firmware file that was emailed?
? Nobody mailed me new firmware...

paulfars
26.01.2018, 21:08
ok this is what they said below.
I have put it on dropbox link https://www.dropbox.com/s/bymn4601ac7nb4l/ProfiLux3_V632.mhx?dl=0
Not had chance to try it myself


Hi


let us first rule out that the P3 firmware included in GCC is different to our final release.

Could you update your P3 with the version attached? Is still 6.32, but maybe the file in GCC was not OK.

Let us go from there. If this was not the reason we must find the differences between your and our system, our system works perfectly.





l

Gael
26.01.2018, 22:22
Hi,

Just received the mail.
But i compared hex of this file and file in gcc and they are same.

gino555
26.01.2018, 22:23
I got sent that email but not had a chance to update the firmware

paulfars
27.01.2018, 18:04
Anybody else got theres working with 6.33 firmware

Gael
27.01.2018, 18:07
I'm not at home to try, i will try 6.33 sunday :-) But i think Matthias found and solved the problem with P3 :-)

paulfars
27.01.2018, 18:08
Yeah I got mine working

Matthias
27.01.2018, 18:31
OK, I will wait with the official release until you three confirmed that it works with P3 now without any issues.

gino555
27.01.2018, 19:02
Will try mine tommorrow

paulfars
27.01.2018, 22:25
Ok , have done some extensive testing tonight , which ranged from being slightly worried to very worried to very pleased !

I wanted to carry on testing until I got two consistent results.

Results below

4.1
5.1
7.3
10.2
12.9
8.4
8.4

So I will take the last two as been accurate as they correspond to my test kit reading.

After the first test the status box showed an error " not enough reagent" even though it showed as been full. I cured this by disabling KH Director and then re enabling.

The only problem I have left is that the record measurement tick box is greyed out.

How quickly will the reagent evaporate ? I have removed the cap and put the tube in the bottle. Do I need to partly seal it?

Matthias
28.01.2018, 09:34
the first 2-3 measurements may be wrong if there has been too much reagent left in the measurement cell after priming the reagent tube
maybe you did not flush the measurement cell after priming (or not enough) or there was just too much reagent in the cell due too much priming

if the values are too high then there was maybe air in the reagent tube, see also hints about the reagent tubing in the manual

when the values settled now in a reasonable area then all seems to be OK


After the first test the status box showed an error " not enough reagent" even though it showed as been full. I cured this by disabling KH Director and then re enabling.
the filling level of the reagent pump must be > 50ml


How quickly will the reagent evaporate ? I have removed the cap and put the tube in the bottle. Do I need to partly seal it?
not so fast, actually I have open bottles all the time here while testing
some use just some tape to seal it

Matthias
28.01.2018, 09:36
The only problem I have left is that the record measurement tick box is greyed out.

measurement recording is supported only for P4 at the moment, it will be added for P3 and GHL Doser 2 with the next updates, I had no chance to get this done until now

Gael
28.01.2018, 18:53
Hi,

It’s work for me too now with 6.33 :-)

Thank

Gael

Matthias
28.01.2018, 19:04
so we take this issue as solved, thanks again for your help

the new GCC 1.1.1.1 has just been uploaded, is available in our download section

Gael
28.01.2018, 20:20
28 January, 8:00 PM:
1st test 8.0dKH
2nd test 9.5dKH
3rd test 9.5dKH
4th test 9.5dKH
5th test 9.5dKH

29 January:
4:00 AM: 9,4dKH
-->5:00 AM +125ml KH <--
7:00 AM: 9,8dKH
1:00 PM: 10,1dKH (why ???)
2:00 PM: Measuring...

gino555
28.01.2018, 21:05
1st test 10.5dKH
2nd test 10.5dKH
3rd test 8.7dKH
4th test 8.7dKH
5th test 8.8dKH

Gael
29.01.2018, 14:20
28 January, 8:00 PM:
1st test 8.0dKH
2nd test 9.5dKH
3rd test 9.5dKH
4th test 9.5dKH
5th test 9.5dKH

29 January:
4:00 AM: 9,4dKH
-->5:00 AM +125ml KH <--
7:00 AM: 9,8dKH
1:00 PM: 10,1dKH (why ???)
2:00 PM: 10,1dKH (Ok, but don't understand how KH increased...)
7:00 PM: 10,0dKH

paulfars
29.01.2018, 15:10
27th Jan
KH dosed at 5.00 PM
KH tested 8.00PM 8.4dKH

28 Jan
KH dosed at 5PM
KH tested 8.00PM 8.7dKH

29th Jan
KH dosed at 5.00AM
KH tested 8.00AM 9.7dKH

I would like to know also why the variation

If i am to allow KH Director to influence the dosing, don't the readings need to be more consistent ?

steviejasp
29.01.2018, 15:33
Mine is the same, All over the place from 3 to 12. I thought it had settled to 6.8 but its now on 5. Must have done about 12 tests now.

Gael
29.01.2018, 15:40
27th Jan
KH dosed at 5.00 PM
KH tested 8.00PM 8.4dKH

28 Jan
KH dosed at 5PM
KH tested 8.00PM 8.7dKH

29th Jan
KH dosed at 5.00AM
KH tested 8.00AM 9.7dKH

I would like to know also why the variation

If i am to allow KH Director to influence the dosing, don't the readings need to be more consistent ?

Maybe you add too many kh?
You test after added KH not before, test also before for help to understand.

Gael

paulfars
29.01.2018, 16:27
will try a test just before the dose is due

Going Deep
17.02.2018, 00:13
I am getting the error: titration taking too long. any help appreciated

paulfars
17.02.2018, 00:19
I am getting the error: titration taking too long. any help appreciated

Have you just setup KHD or was it working ok before the error?

Going Deep
17.02.2018, 02:43
I have just set it up.

paulfars
17.02.2018, 11:25
I have just set it up.

I would double check your setup matches the instructions.

Have you set the dosers to the correct speed and calibrated them. Is the number of the doser configured correctly in KHD

When you click on KHD to take a measurement , what exactly happens?

Going Deep
17.02.2018, 12:28
I was very meticulous and followed the instructions to the T. I tried another measurement and read 11.1 and my Hanna Alk checker read 8.1. Will today if the reading gets better. I am using 50ml sample volume. Should I go back to the 80ml default?

paulfars
17.02.2018, 12:39
I was very meticulous and followed the instructions to the T. I tried another measurement and read 11.1 and my Hanna Alk checker read 8.1. Will today if the reading gets better. I am using 50ml sample volume. Should I go back to the 80ml default?

Mine has always been set to default 80ml so I can't comment on how it would effect the readings. I had to do about 10 measurements before it settled down to give constant readings.

Going Deep
18.02.2018, 11:44
Thanks for the advice. I have been sampling for a day now and have been consistently getting measurements around 6.6 (four samples). But my corresponding Hanna Checker reading is 7.8. Is this too big a discrepancy? What would cause a low measurement. My sample volume is 80ml.

paulfars
18.02.2018, 17:55
Thanks for the advice. I have been sampling for a day now and have been consistently getting measurements around 6.6 (four samples). But my corresponding Hanna Checker reading is 7.8. Is this too big a discrepancy? What would cause a low measurement. My sample volume is 80ml.

Make sure there is no air bubbles in the reagent line and also make sure that the vent pipe is not submerged in water so it can't suck water back into KHD.

Are you sure that your Hanna reading is accurate?

gino555
18.02.2018, 20:20
I checked my salifert test kit and it was reading .7kh high


might be worth getting a reference solution to check accuracy of the Hanna

Going Deep
19.02.2018, 12:30
I replaced my Hanna Reagent and is now consistent with my KHD measurement. I am also ordering a backup alk measurement for redundancy. Keep fingers crossed.....right now I am loving KHD directxor!