PDA

View Full Version : Cannot connect 2nd PH probe



subshop
12.03.2009, 04:32
I connected the probe to the Profilux. Unplugged the plugged the Profilux back in but I do not see it reading the second probe. Thank you in advance, Neil

GHL Support
12.03.2009, 08:45
Do you have the firmware 4.04 or higher?
Which ProfiLux type do you have?
Have you made the corresponding settings after having installed the extension card?

Matthias
12.03.2009, 09:09
Don`t forget to enable the second pH-probe under the "extras->display" settings.

subshop
12.03.2009, 14:41
Don`t forget to enable the second pH-probe under the "extras->display" settings.

Matthias, I tried that and there was no setting for PH probe 2. Also checked under "system", in the software controller, and it only shows sensors 1-4:
PH-sensor 1
Temp-sensor 2
Conductivity- sensor 3
Redox-sensor 4

Could it be that the Profilux is not reading the Ph card?
Neil

subshop
12.03.2009, 14:51
oh, i also forgot to mention this has happened with 4.04 and 4.05 firmware. Neil

Matthias
12.03.2009, 15:18
are we talking about a ProfiLux Plus II eX?
If yes - you can`t use slot 0 for measurement cards - this is mentioned in the manual.
Please use another slot.

subshop
12.03.2009, 15:51
are we talking about a ProfiLux Plus II eX?
If yes - you can`t use slot 0 for measurement cards - this is mentioned in the manual.
Please use another slot.

Matthias, yes it is. I did not see that in the manual but will move to a different slot and try again. thank you, Neil

subshop
13.03.2009, 16:06
Matthias, as usual you were correct. Moved card to a different slot and it works great. Thanks again, neil